[Eeglablist] EEG amplifier and alternative input
Frank H Duffy
fhd at sover.net
Wed Aug 4 08:16:31 PDT 2010
There are two major issues, both related to the differences in EEG and
voice spectral content:
1. EEG amplifiers typically record from 0.5 -100 Hz but may be able to
open to 500Hz for EP data collection. However, voice typically requires,
for telephone-like reproduction, a 500 Hz to 3000 Hz band pass. Thus the
EEG amplifiers would likely reject and/or severly distort audio inputs.
2. As most EEG machines are digital these days, they time sample
(digitize) EEG signals. As a minimum they must use a rate that is
slightly more than twice the highest desired input signal, e.g., 200 H
samplige rate for 100 Hz EEG. For voice one would require a
digitization rate of at least 2x 3000Hz; typically one wants a minimum
of 10000 Hz for voice. So the EEG devices typical sampling rates of
200-1024 Hz would not reasonably sample voice signals.
So, no, voice over EEG amplifiers won't work. It would take a complex,
parallel EEG and separate voice amplifier and A-D converter setups to
match voice to EEG. This can be done. Indeed some EEG machines match
both voice and video to EEG. However, there is a lot of engineering to
make the EEG and voice/video follow the same clock time.
Best of luck.
Frank Duffy, MD fhd at sover.net
For the above discussion also BSEE and ham (k1moq)
On 8/3/2010 8:10 AM, Marc wrote:
> Hi.
>
> This is not directly related to EEGLab. But I thought someone may be
> able to give some pointers.
>
> We're doing some experiments measuring scalp EEG of participants. One
> of the variables we intend to record is the participant's voice reply.
> We're wondering if it is possible to use the same EEG hardware
> amplifier to record the voice reply? Have anyone tried that before?
> That is, instead of connecting the electrodes to the one of the 64
> channels of the amplifier, we connect a microphone to the input
> channel. I assume we will be able to record as good a signal as any
> audio amplifier? Is there any thing else to watch out for?
>
> Thanks for any advise.
>
> Marc.
> _______________________________________________
> Eeglablist page: http://sccn.ucsd.edu/eeglab/eeglabmail.html
> To unsubscribe, send an empty email to eeglablist-unsubscribe at sccn.ucsd.edu
> For digest mode, send an email with the subject "set digest mime" to eeglablist-request at sccn.ucsd.edu
>
>
More information about the eeglablist
mailing list