Thank you for responding!
Comparing my old settings to new correctly filtered ones there do not seem
to be notable differences, so I will not panic for the time being.
I look forward to the upcoming version of pop_firws. Will that be coming
with eeglab 14?
After reading your article, it seems like larger transition bandwidths and
lower filter orders provide less-distorting roll off of the data.
The options I mentioned previously a .02Hz bandwidth for the highpass and
22Hz for the lowpass. In your example you listed 10Hz for the lowpass. I am
wondering if I should use the shorter 10Hz, although again my own
comparisons do not show large differences between the two. I do not want to
filter too sharply.
Thank you,
2017-02-28 5:01 GMT-05:00 Andreas Widmann <widmann at uni-leipzig.de>:
> Hi Isaiah,
>
> > I have been closely following the recent pop_eegfiltnew filtering
> discussion with great interest; I would like to verify that I am filtering
> my datasets correctly. My goal is to bandpass filter 0.1 - 100Hz
> >
> > Previously I had been using these options:
> > EEG = pop_eegfiltnew(EEG, 0.1, 100, 33000, 0, [], 0);
> > I believe I understand now that this is wrong because 1) I should not
> use the highpass filter order for the entire bandpass, but do them
> separately and 2) the input is passband edge, NOT cutoff frequency. In
> reality here I was filtering from 0.05 - 112.5Hz correct?
> No, cutoff frequencies are 0.05 and 100.05. Transition bandwidth is 0.1
> here. The cutoff frequency is passband edge +/- half the transition
> bandwidth.
>
> Compute transition band width df as ( dF / order ) * fs. dF is a constant
> defined by the window type. pop_eegfiltnew uses Hamming window with dF =
> 3.3. Thus,
> df = ( 3.3 / 33000 ) * 1000 = 0.1
>
> > How worried should I be about my mistake?
> Depends. Usually not too much, but if you have stronger low frequency
> noise or drifts or sharp artifacts or signal discontinuities these setting
> could also be problematic. I would recommend following your approach below.
>
> > To correctly filter 0.1-100Hz (at 1kHz srate) I should:
> > EEG = pop_eegfiltnew(EEG, 0.2, [], 16500, 0, [], 0);
> > EEG = pop_eegfiltnew(EEG, [], 89, 150, 0, [], 0);
> > Is this accurate?
> Yes.
>
> > I have looked at using pop_firws, but am hesitant because it does not
> display in the command window the cutoff freq, passband edge, etc…
> This is already added in the upcoming version. Will take some days/weeks
> still.
>
> > and I am not certain how to define the transition bandwidth correctly
> here.
> Transition bandwidth is defined by filter order. You can use the inverted
> equation from above or much easier press the „Estimate“ button, enter your
> requested transition bandwidth and get back the required filter order
> already filled in. For example, if want a lowpass cutoff of 100 Hz with a
> passband edge at 95 Hz you need a transition band width of 10 Hz (see
> above). Enter: cutoff 100 Hz, lowpass, Hamming window, click Estimate,
> enter transition bandwidth 10 Hz, click Ok, required filter order 330 (m =
> dF / ( df / fs ) = 3.3 / ( 10 / 1000 ) = 330) is already filled, click Plot
> filter responses to check whether the filter is designed as expected and
> apply with ok.
>
> I recommend reading
> http://home.uni-leipzig.de/~biocog/eprints/widmann_
> a2015jneuroscimeth250_34.pdf
> for a hopefully accessible introduction into the basic concepts. No math
> or magic required :)
>
> Best,
> Andreas
>
> >
> > I would like to thank Andreas, Jumana, Makoto, and any others for
> having this discussion; it has been helpful.
> >
> >
> > Thank you,
> > ---------- Mensaje reenviado ----------
> > From: Makoto Miyakoshi <mmiyakoshi at ucsd.edu>
> > To: Andreas Widmann <widmann at uni-leipzig.de>
> > Cc: EEGLAB List <eeglablist at sccn.ucsd.edu>
> > Date: Tue, 7 Feb 2017 19:05:31 -0800
> > Subject: Re: [Eeglablist] high frequency oscillation- eeg advice
> > Dear Andreas and Jumana,
> >
> > > (0.2 - 0.2 / 2 and 35 + 10 / 2, respectively).
> >
> > This is much more understandable for I could not remember how
> eegfiltnew() determines the transient band width for the low-pass end.
> >
> > So for the higher frequencies (above 10Hz), log scale (octaves) is not
> used to determine the transient band width. I guess this is related to
> conventional practice in EEG research.
> >
> > Makoto
> >
> >
> > On Fri, Feb 3, 2017 at 5:26 PM, Andreas Widmann <widmann at uni-leipzig.de>
> wrote:
> > I hope it is ok to keep the discussion on-list. It might be interesting
> also for others.
> >
> > > Thanks again. I wanted to follow Steve Luck's ERP advice and filter
> between 0.1 as a half amplitude cutoff - 30/40Hz. I went for 40. What does
> this mean in terms of EEGlab's passband edge?
> > You only have to shift the passband edge by half the transition band
> width. In your case I would double the transition band width for the
> high-pass part to reduce the order.
> >
> > if EEG.srate == 256;
> > EEG = pop_eegfiltnew(EEG, 0.2,[], 4224, 0, [], 1);
> > EEG = pop_eegfiltnew(EEG, [], 35, 86, 0, [], 0);
> > elseif EEG.srate == 1000;
> > EEG = pop_eegfiltnew(EEG, 0.2,[], 16500, 0, [], 0);
> > EEG = pop_eegfiltnew(EEG, [], 35, 330, 0, [], 0);
> > end
> >
> > will give you 0.1 Hz high-pass and 40 Hz low-pass cutoff frequencies
> (0.2 - 0.2 / 2 and 35 + 10 / 2, respectively). In general it is easier to
> directly use the windowed sinc filter function to specify the cutoff (but
> will give the identical result here if used with Hamming window).
> >
> > > I also wanted to follow Makoto's advice of using 1 Hz filter for
> running ICA (I apply the weights to 0.1). Was I therefore incorrect to use
> 1-40Hz as a passband edge?
> > I assume this also referred to cutoff frequency rather than passband
> edge (similar as in the Winkler et al., 2015, publication). However, I
> think the impact is limited here.
> >
> > Best,
> > Andreas
> >
> > > Best wishes,
> > > Jumana
> > >
> > > -----Original Message-----
> > > From: Andreas Widmann [mailto:widmann at uni-leipzig.de]
> > > Sent: 03 February 2017 15:03
> > > To: Ahmad, Jumana <jumana.ahmad at kcl.ac.uk>
> > > Cc: mmiyakoshi at ucsd.edu; eeglablist at sccn.ucsd.edu
> > > Subject: Re: [Eeglablist] high frequency oscillation- eeg advice
> > >
> > > Filters need transition bands separating passband and stopband. For
> FIR filters the cutoff frequency is in the center of the transition band.
> In pop_eegfiltnew you specify passband edges but reported are most commonly
> the cutoff frequencies. You compute cutoff frequency as passband edge +/-
> transition band width / 2, here 0.1 - 0.1 / 2 and 40 + 10 / 2. Transition
> band width is reported in the console output of pop_eegfiltnew and can also
> be computed manually from order. See
> > > https://emea01.safelinks.protection.outlook.com/?url=
> http:%2F%2Fhome.uni-leipzig.de%2F~biocog%2Feprints%2Fwidmann_
> a2015jneuroscimeth250_34.pdf&data=01%7C01%7Cjumana.ahmad%40kcl.ac.uk%
> 7Cf50ed9fcf01b42b49aee08d44c45bef8%7C8370cf1416f34c16b83c72407165
> 4356%7C0&sdata=GxppGkTzfuGeht3gOpalHbv2dG5%2FhJwmOJFD5%2BCaj8M%3D&
> reserved=0
> > > for an introduction into the basic concepts (and also the equations to
> compute transition band width and order).
> > >
> > > Hope this helps,
> > > Andreas
> > >
> > >> Am 03.02.2017 um 15:49 schrieb Ahmad, Jumana <jumana.ahmad at kcl.ac.uk
> >:
> > >>
> > >> Would it be 40Hz low pass? I am unsure why 0.1 has been halved but 40
> hasn't?
> > >> Thank you,
> > >> Jumana
> > >>
> > >>
> > >> -----Original Message-----
> > >> From: Andreas Widmann [mailto:widmann at uni-leipzig.de]
> > >> Sent: 03 February 2017 14:45
> > >> To: Ahmad, Jumana <jumana.ahmad at kcl.ac.uk>
> > >> Cc: mmiyakoshi at ucsd.edu; eeglablist at sccn.ucsd.edu
> > >> Subject: Re: [Eeglablist] high frequency oscillation- eeg advice
> > >>
> > >> Yes, that looks ok. Be aware that passband edge and not cutoff
> frequency is specified in pop_eegfiltnew (for backward compatibility).
> Report these filters as 0.05 Hz high-pass and 45 Hz low-pass (incl. specs:
> zero-phase Hamming windowed sinc FIR, order, and optimally transition band
> width).
> > >>
> > >> Best,
> > >> Andreas
> > >>
> > >>> Am 02.02.2017 um 13:00 schrieb Ahmad, Jumana <jumana.ahmad at kcl.ac.uk
> >:
> > >>>
> > >>> Thank you both for your explanation.
> > >>>
> > >>> Does this therefore look reasonable:
> > >>>
> > >>> if EEG.srate == 256;
> > >>>
> > >>> EEG = pop_eegfiltnew(EEG, 0.1,[], 8448, 0, [], 1);
> > >>> EEG = pop_eegfiltnew(EEG, [], 40, 86, 0, [], 0);
> > >>>
> > >>> elseif EEG.srate == 1000;
> > >>> EEG = pop_eegfiltnew(EEG, 0.1,[], 33000, 0, [], 0);
> > >>> EEG = pop_eegfiltnew(EEG, [], 40, 330, 0, [], 0);
> > >>> end
> > >>>
> > >>> I have different sample rates. I work with a very large dataset
> recorded with different systems.
> > >>> Best wishes,
> > >>> Jumana
> > >>>
> > >>> From: eeglablist-bounces at sccn.ucsd.edu
> > >>> [mailto:eeglablist-bounces at sccn.ucsd.edu] On Behalf Of Makoto
> > >>> Miyakoshi
> > >>> Sent: 31 January 2017 03:29
> > >>> To: Andreas Widmann <widmann at uni-leipzig.de>
> > >>> Cc: eeglablist at sccn.ucsd.edu
> > >>> Subject: Re: [Eeglablist] high frequency oscillation- eeg advice
> > >>>
> > >>> Dear Andreas,
> > >>>
> > >>> Thank you for your explanations.
> > >>>
> > >>>> This was just a wild guess based on experience.
> > >>>
> > >>> Oh ok I see.
> > >>>
> > >>>> But note that the EEGLAB iirfilt-plugin does implement an Elliptic
> and not a Butterworth IIR filter.
> > >>>
> > >>> That's true. If I remember correctly, Elliptic filter can make
> theoretically the sharpest filter.
> > >>>
> > >>>> In MATLAB you compute the impulse response with impz (see the code
> example in my previous post).
> > >>>
> > >>> Yes, this is a new thing I learned this time! This function is very
> convenient.
> > >>>
> > >>> Thanks Andreas as always.
> > >>>
> > >>> Makoto
> > >>>
> > >>>
> > >>>
> > >>> On Sat, Jan 28, 2017 at 1:26 AM, Andreas Widmann <
> widmann at uni-leipzig.de> wrote:
> > >>> Dear Makoto
> > >>>
> > >>>>>> A roughly equivalent Butterworth filter will have an estimated
> effective (numerically relevant) impulse response duration of 20592 samples
> (~82s!):
> > >>>>
> > >>>> I tried to replicate it with EEGLAB iirfilt() but the filter order
> was only 6. How did you calculate this number?
> > >>> Not sure whether I understand the question.
> > >>>
> > >>> Do you mean how I estimated the order (4) of the Butterworth filter
> in my example to be "roughly equivalent“ to the FIR variant? This was just
> a wild guess based on experience. In case necessary I could check how rough
> "roughly equivalent" really is. But note that the EEGLAB iirfilt-plugin
> does implement an Elliptic and not a Butterworth IIR filter.
> > >>>
> > >>> Or do you mean the comparison of your iirfilt order (6) to the
> impulse response duration (20592) of my example? One cannot directly
> compare the order of IIR and FIR filters. This is a (unfortunately common)
> misconception. For FIR filters (implemented by convolution) the order
> equals the duration of the impulse response (minus one sample). This does
> not hold for IIR filters implemented *recursively*. However, even if the
> impulse response of IIR filters is infinite by definition in digital
> implementations it is practically limited by numerical precision. That is,
> you can compute the impulse response of a digital IIR filter which is now
> also finite and can be applied by convolution exactly like a FIR filter
> finally giving the same result (within limits of numerical precision).
> > >>>
> > >>> In MATLAB you compute the impulse response with impz (see the code
> example in my previous post). If I remember correctly you may specify
> required numerical precision as an input argument to impz. I trusted the
> default here. You have to get the filter coefficients (b, a) of your EEGLAB
> iirfilt filter and feed them to impz. The length of the resulting impulse
> response times two is the duration to be compared to the order (or IR
> duration) of a FIR filter. Times two because you have to apply the
> non-linear phase IIR derived filter in forward and reverse direction (to
> get zero-phase).
> > >>>
> > >>> Does this help? Best,
> > >>> Andreas
> > >>>
> > >>>>
> > >>>> Makoto
> > >>>>
> > >>>>
> > >>>>
> > >>>> On Fri, Jan 27, 2017 at 2:35 PM, Andreas Widmann <
> widmann at uni-leipzig.de> wrote:
> > >>>> Dear Makoto,
> > >>>>
> > >>>> to my understanding these are two different and mostly unrelated
> issues:
> > >>>>
> > >>>> (1) Recommendations for very low high-pass cutoff frequencies:
> > >>>> The filter implementation (IIR vs. FIR) doesn’t really matter here.
> It is correct that you need very long epochs to resolve these very low
> frequencies. However, it is not valid to reverse this conclusion. Also, for
> short epochs apparently small differences in cutoff frequency may matter.
> You do filter the continuous data, that is, epoch length does not directly
> influence the filter effects. High-pass filter distortions (mainly due to
> larger late and slow components biasing smaller early and fast components)
> were shown empirically several times, e.g. most convincingly by Acunzo et
> al.
> > >>>>
> > >>>> But I agree that this issue is a matter of debate (mainly affecting
> ERP research). Personally, I recommend to individually adjust filter
> cutoffs to the properties of signal and noise. It is in principle possible
> to apply higher high-pass cutoff frequencies if noise level really requires
> this and it is properly demonstrated that filter distortions do not bias
> the conclusions. But people always ask for general recommendations and you
> have to be on the save side...
> > >>>>
> > >>>> (2) IIR vs. FIR
> > >>>> In the first part you argue against long impulse response
> durations. For IIR filters these are usually even (considerably) longer.
> Besides other issues (as the non-linear phase property of IIR filters and
> its side-effects) this is the main reason why I do not recommend IIR
> filters for offline processing in electrophysiology/ERPs. Following your
> example with a slightly higher cutoff of 0.1 Hz and a Hamming windowed sinc
> FIR filter you will have a impulse response duration of 4127 samples
> (~16s). A roughly equivalent Butterworth filter will have an estimated
> effective (numerically relevant) impulse response duration of 20592 samples
> (~82s!):
> > >>>>
> > >>>> [b, a]=butter(4, 0.1/125, 'high');
> > >>>> h = impz(b, a);
> > >>>> length(h)*2 % Times 2 as you have to apply the non-linear phase
> > >>>> filter in forward and reverse direction to achieve zero-phase
> > >>>> doubling the impulse response duration
> > >>>>
> > >>>> Thus, if you define filter efficiency as achieved roll-off per
> impulse response duration (as I do) IIR filters can be considerably less
> efficient. Impulse response duration defines the extent the signal is
> convoluted and artifacts and distortions are potentially smeared.
> > >>>>
> > >>>>> 0.1Hz IIR high-pass filter can be found in EEG amplifiers, so in
> that sense it is not strange at all.
> > >>>>
> > >>>> IIR filters have their merits if high throughput and/or low delays
> are required. FIR filters have large delays and are computationally less
> efficient (there are however highly optimized implementations). Delay and
> throughput do not matter for offline processing but both are highly
> relevant in amplifiers and BCI applications. Thus, IIR filters are
> frequently preferred for online processing (despite non-linear phase).
> > >>>>
> > >>>>> I thought it seemed a bit odd to report it like this in a paper.
> > >>>>
> > >>>> I do not see a problem as long as you reason your approach.
> > >>>>
> > >>>> Best,
> > >>>> Andreas
> > >>>>
> > >>>>> Am 27.01.2017 um 21:34 schrieb Makoto Miyakoshi <
> mmiyakoshi at ucsd.edu>:
> > >>>>>
> > >>>>> Dear Andreas,
> > >>>>>
> > >>>>>>> And I still don't like 0.1Hz high-pass if you use FIR
> > >>>>>> Why? What is the problem?
> > >>>>>
> > >>>>> Assuming people just enter '0.1' to EEGLAB default FIR filter GUI,
> it'll apply 0.05Hz cut-off high-pass filter with Hamming window. When the
> sampling rate is 250Hz, the model order it calculates is 8251, which is 33
> sec long. Meanwhile, people are usually only interested in the first few
> hundreds milliseconds of the averaged signal. And they subtract
> pre-stimulus baseline mean value from the entire epoch anyway. I don't see
> much reason to apply 0.1-Hz high-pass filter in these cases. Of course, I
> saw papers discussing this issue and I have no objection, but intuitively
> it is still weird to me. I think the point is that people want to claim
> that ERP is a broadband phenomenon, but the way they demonstrate it is not
> satisfactory to me.
> > >>>>>
> > >>>>> By the way, people often complain about our recommended -1 to 2
> sec epoch to be too long, and our recommended 1-Hz high-pass filter too
> aggressive. But doesn't it makes more sense to apply 1-Hz high-pass filter
> to 3-sec epoch data, compared with applying 0.1-Hz high-pass for 0.8 sec
> epoch data? This kind of unbalancedness makes me feel weird.
> > >>>>>
> > >>>>>> What would be your suggested alternative?
> > >>>>>
> > >>>>> I thought IIR would be more reasonable for such a low cutoff
> frequency, but I have never tried it myself (as far as I know,
> clean_rawdata plugin comes with its IIR filter, which could be a part of
> BCILAB for online processing). It depends on your relative time scale. If I
> analyze hour-long resting state data to target minute-long slow changes, I
> would use FIR with no problem.
> > >>>>>
> > >>>>> I did not know much about stability issue, but this time you made
> me learn it a little bit. Thank you Andreas for always pushing my back in
> this way.
> > >>>>>
> > >>>>> Makoto
> > >>>>>
> > >>>>>
> > >>>>>
> > >>>>> On Fri, Jan 27, 2017 at 3:28 AM, Andreas Widmann <
> widmann at uni-leipzig.de> wrote:
> > >>>>> Dear Makoto,
> > >>>>>
> > >>>>>> And I still don't like 0.1Hz high-pass if you use FIR
> > >>>>> Why? What is the problem? What would be your suggested alternative?
> > >>>>>
> > >>>>>> (and I do not know how bad it is to use IIR... I've heard it can
> > >>>>>> become 'unstable' but I've never seen it myself)
> > >>>>> Here you go:
> > >>>>> [b,a]=butter(4,0.1/500,'high');
> > >>>>> isstable(b,a)
> > >>>>> freqz(b,a)
> > >>>>>
> > >>>>> But note that possible instability is not the main problem with
> IIR application in electrophysiology. There are workarounds (e.g. for this
> example using zpk: [z,p,k]=butter(4,0.1/500,'high'); sos=zp2sos(z,p,k);
> isstable(sos)).
> > >>>>>
> > >>>>> Best,
> > >>>>> Andreas
> > >>>>>
> > >>>>>
> > >>>>>>
> > >>>>>>> I won't be using granger causality but I will be estimating
> phase during ITC.
> > >>>>>>
> > >>>>>> Should be ok.
> > >>>>>>
> > >>>>>> Makoto
> > >>>>>>
> > >>>>>> On Thu, Jan 26, 2017 at 2:12 PM, Ahmad, Jumana <
> jumana.ahmad at kcl.ac.uk> wrote:
> > >>>>>> Dear Makoto,
> > >>>>>> I actually switched to the pop eeg filt eeglab function and it
> now Really attenuated anything >40Hz, and my ERPs are cleaner. However I
> filtering between 0.1-40Hz at the same time in the GUI (I interned the high
> pass and low pass simultaneously). Is this OK to do? The frequency response
> looks OK.
> > >>>>>>
> > >>>>>> The filter order was automatically set very high by the GUI, but
> it's continuous data and I have room without events at the beginning and
> end of the data so any edge effects can be disgusted. What do you think?
> > >>>>>>
> > >>>>>> Also, this is for my ERP analysis- I trained ICA on a 1Hz high
> pass filtered set.
> > >>>>>> I won't be using granger causality but I will be estimating phase
> during ITC.
> > >>>>>> Best wishes, and thanks,
> > >>>>>> Jumana
> > >>>>>>
> > >>>>>> ------------------------------------------
> > >>>>>> Jumana Ahmad
> > >>>>>> Post-Doctoral Research Worker in Cognitive Neuroscience EU-AIMS
> > >>>>>> Longitudinal European Autism Project (LEAP) & SynaG Study Room
> > >>>>>> M1.26.Department of Forensic and Neurodevelopmental Sciences (PO
> > >>>>>> 23) | Institute of Psychiatry, Psychology & Neuroscience | King’s
> > >>>>>> College London | 16 De Crespigny Park | London SE5 8AF
> > >>>>>>
> > >>>>>> Phone: 0207 848 5359| Email: jumana.ahmad at kcl.ac.uk | Website:
> > >>>>>> https://emea01.safelinks.protection.outlook.com/?url=www.eu-aims
> > >>>>>> .eu&data=01%7C01%7Cjumana.ahmad%40kcl.ac.uk%7C15add3a4fe3049459e
> > >>>>>> 8a08d44c434afd%7C8370cf1416f34c16b83c724071654356%7C0&sdata=9bFL
> > >>>>>> Fj%2FkVl61CL%2BK749qenX14JOTzEBSNn9t9%2BBCJUk%3D&reserved=0 |
> > >>>>>> Facebook:
> > >>>>>> https://emea01.safelinks.protection.outlook.com/?url=www.faceboo
> > >>>>>> k.com%2Feuaims&data=01%7C01%7Cjumana.ahmad%40kcl.ac.uk%7C15add3a
> > >>>>>> 4fe3049459e8a08d44c434afd%7C8370cf1416f34c16b83c724071654356%7C0
> > >>>>>> &sdata=d3Z6OAcHSkY4ww7a%2Bda2ir1pLENlcVPz%2FsDgLl%2F51fo%3D&rese
> > >>>>>> rved=0
> > >>>>>>
> > >>>>>> From: Makoto Miyakoshi <mmiyakoshi at ucsd.edu>
> > >>>>>> Sent: 26 January 2017 23:49:50
> > >>>>>> To: Ahmad, Jumana
> > >>>>>> Cc: eeglablist at sccn.ucsd.edu
> > >>>>>> Subject: Re: [Eeglablist] high frequency oscillation- eeg advice
> > >>>>>>
> > >>>>>> Dear Jumana,
> > >>>>>>
> > >>>>>> It's a bad idea to perform ICA with 0.1Hz high-pass filtered
> data. The cutoff frequency is too low. See this page and the referenced
> paper.
> > >>>>>>
> > >>>>>> https://emea01.safelinks.protection.outlook.com/?url=https%3A%2F
> > >>>>>> %2Fsccn.ucsd.edu%2Fwiki%2FMakoto%2527s_preprocessing_pipeline%23
> > >>>>>> High-pass_filter_the_data_at_1-Hz_.28for_ICA.2C_ASR.2C_and_Clean
> > >>>>>> Line.29&data=01%7C01%7Cjumana.ahmad%40kcl.ac.uk%7C15add3a4fe3049
> > >>>>>> 459e8a08d44c434afd%7C8370cf1416f34c16b83c724071654356%7C0&sdata=
> > >>>>>> N0z4Ru6QsY24Bg4PD7FFPrfC999R8%2F7hLVk5ZvonLDU%3D&reserved=0
> > >>>>>>
> > >>>>>>> A 30Hz low pass does not help to get rid of the oscillation,
> which is really significant in the data.
> > >>>>>>
> > >>>>>> Check the channel frequency spectra and tell me if you see peaks
> > >>>>>> in it. If necessary, you can cut it off using a designed low-pass
> > >>>>>> filter (not like Butterworth...)
> > >>>>>>
> > >>>>>>> I use a butterworth filter, which is good for ERP analysis with
> low phase distortion.
> > >>>>>>
> > >>>>>> Do not make qualitative judgement just because something is NOT a
> classic Butterworth. Of course, if the attenuation is small, the phase
> 'distortion' is small. But if such small attenuation is not useful, it does
> not help at all! Also, be careful with the word 'phase'. Particularly
> people who do not know basics of signal processing believe phase as some
> magical thing. If you are not performing Granger Causality Analysis or
> something, you don't need to be so worried about phase issue in practice.
> > >>>>>>
> > >>>>>>> I also already run ICA, but in some datasets there is a very
> significant high frequency oscillation.
> > >>>>>>
> > >>>>>>
> > >>>>>> Remember, to eliminate this is more important than being afraid
> of qualitative phase issue.
> > >>>>>>
> > >>>>>>> However, I can see the high frequency oscillations in my ERP,
> which is not ideal and now I need to try and get rid of it further.
> > >>>>>>
> > >>>>>> Can I filter again on top of the data which already has already
> undergone ICA- I only use ICA to remove blinks?
> > >>>>>>
> > >>>>>>
> > >>>>>> You'd better to filter the data on continuous state. If you need
> to filter the epoched data, the half of filter length from both ends
> becomes unreliable.
> > >>>>>>
> > >>>>>>> Should I do cleanline, although it would have to be after ICA
> now- I read this is not advisable.
> > >>>>>>
> > >>>>>>> Should I use a notch filter?
> > >>>>>>
> > >>>>>>
> > >>>>>> If you see > 20dB line noise, Cleanline may not help. In this
> case, I would simply apply a designed low-pass filter, either Hamming
> (-50dB) or Blackman (-70dB) using firfilt(). See 'Tools' -> 'Filter the
> data' -> 'Windowed sinc FIR filter'.
> > >>>>>>
> > >>>>>> There are different guys saying different things about data
> preprocessing. It is confusing, I know! The only good solution for this is
> to become an engineer yourself...
> > >>>>>>
> > >>>>>> Makoto
> > >>>>>>
> > >>>>>>
> > >>>>>>
> > >>>>>> On Tue, Jan 24, 2017 at 7:00 AM, Ahmad, Jumana <
> jumana.ahmad at kcl.ac.uk> wrote:
> > >>>>>> Hi Everyone,
> > >>>>>>
> > >>>>>> I am running a large scale ERP analysis. I filtered 1-40Hz (ICA
> AMICA), or 0.1-40Hz for the ERP dataset. A 30Hz low pass does not help to
> get rid of the oscillation, which is really significant in the data. I use
> a butterworth filter, which is good for ERP analysis with low phase
> distortion.
> > >>>>>>
> > >>>>>> I also already run ICA, but in some datasets there is a very
> significant high frequency oscillation.
> > >>>>>>
> > >>>>>> I do not use cleanline, which is not typical in the literature I
> have been basing my pipeline on.
> > >>>>>>
> > >>>>>>
> > >>>>>>
> > >>>>>> However, I can see the high frequency oscillations in my ERP,
> which is not ideal and now I need to try and get rid of it further.
> > >>>>>>
> > >>>>>> Can I filter again on top of the data which already has already
> undergone ICA- I only use ICA to remove blinks?
> > >>>>>>
> > >>>>>> Should I do cleanline, although it would have to be after ICA
> now- I read this is not advisable.
> > >>>>>>
> > >>>>>> Should I use a notch filter?
> > >>>>>>
> > >>>>>>
> > >>>>>>
> > >>>>>> Any help would be appreciated.
> > >>>>>>
> > >>>>>> Best wishes,
> > >>>>>>
> > >>>>>> Jumana
> > >>>>>>
> > >>>>>>
> > >>>>>>
> >
> > --
> > Isaiah Innis
> > Indiana University '13
> > EEG Technician, IUB IRF
> >
> > _______________________________________________
> > Eeglablist page: http://sccn.ucsd.edu/eeglab/eeglabmail.html
> > To unsubscribe, send an empty email to eeglablist-unsubscribe at sccn.
> ucsd.edu
> > For digest mode, send an email with the subject "set digest mime" to
> eeglablist-request at sccn.ucsd.edu
>
>
--
Isaiah Innis
Indiana University '13
EEG Technician, IUB IRF
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://sccn.ucsd.edu/pipermail/eeglablist/attachments/20170302/719030fd/attachment.html>
More information about the eeglablist