In EEG recordings do we expect movement artifact (due to liquid junction
potential) to be expressed at slow frequencies or sudden potential
difference spikes (i.e. temporary electrical discontinuities)? I have found
empirical evidence of both.
Example references:
Isolating gait-related movement artifacts in electroencephalography during
human walking
<https://www.researchgate.net/profile/Kristine_Snyder/publication/278792694_Isolating_gait-related_movement_artifacts_in_electroencephalography_during_human_walking/links/55873f3e08ae71f6ba914812.pdf>
Electroencephalography (EEG): Neurophysics, Experimental Methods, and
Signal Processing
<https://www.researchgate.net/profile/Michael_Nunez4/publication/290449135_Electroencephalography_EEG_neurophysics_experimental_methods_and_signal_processing/links/57bf32c908ae2f5eb32e82a9/Electroencephalography-EEG-neurophysics-experimental-methods-and-signal-processing.pdf>
On Thu, May 3, 2018 at 12:20 PM, Makoto Miyakoshi <mmiyakoshi at ucsd.edu>
wrote:
> Dear colleagues,
>
> Let me share this info.
>
> E. Huigen (2000) Noise in biopotential recording using surface electrodes
> https://www.semanticscholar.org/paper/Noise-in-
> biopotential-recording-using-surface-Huigen/8fc0837f7af0a36b19799139d9b763
> 969057b985
>
> *%%%%%%%%%%%%%%%%%%%%%*
>
> *2.4 Motion artifact*
> Movement can cause changes in the potentials that are created when an
> electrode is applied to the skin. Normally, when the patient is relaxed,
> and high quality electrodes are used, the recording is not distorted by
> motion artifact. Brinkman (1993) has found no significant correlation
> between the intentional movement of the arm and the noise signal. The
> mechanisms that can cause motion artifact are described next.
>
> *Liquid junction potential variations by electrode movement*
> The various phase junctions in the electrode-electrolyte-skin interface
> all cause junction potentials, sensitive to motion artifact. The
> skin-electrolyte interface can cause artifacts of 400-600 μV when the
> electrode is moved parallel to the skin surface (Smith and Wace, 1995).
> When the electrode is moved perpendicular to the skin the potential changes
> can be up to 900 μV. Firm attachment to the skin can reduce the potential
> changes. The electrode-electrolyte interface also produces artifacts when
> mechanically disturbed. Gatzke (1974, as described in Webster, 1984)
> measured a 15 mV potential change when a pure silver electrode is moved
> in electrolyte. Coating with silver chloride, thus creating a stable double
> layer, produces a 10-fold reduction of the artifact. Further reduction (up
> to negligible value) can be achieved by recessing the electrode-electrolyte
> interface in a protective cup, in which a sponge soaked in gel is placed
> (figure 2-2).
>
> *Skin potential changes*
> Earlier, the stratum corneum and the barrier layer have been identified as
> the major sources of the impedance of the skin. Webster (1984) has also
> observed a potential difference between the inside and the outside of the
> barrier layer. Van Wijk van Brievingh (1988) however states that this
> skin potential is a liquid junction potential between deeper skin layers
> and the electrolyte. The skin potential has a typical value of 30 mV at
> the thorax. Stretching of the skin decreases the skin potential to about 25
> mV. This artifact can be reduced to negligible value by abrading the skin.
> Webster gives 20 strokes with fine sandpaper as an indication. Shackel
> (1959, as described in Geddes and Baker, 1989) developed a method for
> shorting the skin potential, called the skin-drilling technique. The
> epidermis is eroded using a dental burr. The capillaries remain
> undisturbed, so no blood is drawn. Unfortunately, the epidermis is also the
> layer that protects the skin from irritation. A mild electrode gel has to
> be used to prevent
> unwanted effects.
>
> %%%%%%%%%%%%%%%%%%%%%
>
> Wikipedia 'liquid junction potential'
> https://en.wikipedia.org/wiki/Liquid_junction_potential
>
> In Huigen, Peper, Grimbergen (2002) Med. biol. Eng, they described:
>
> *RECORDINGS OF biomedical signals from the body surface often
> contain a substantial noise component. This noise signal can
> severely impair the resolution of biomedical recordings as its
> magnitude can be as high as 10-60 uVp_p (GONDRAN et aL, 1996).
> This is in the range of EEG potentials and is at least ten
> times as high as several types of evoked potentials (e.g. visual
> evoked potentials or somatosensory evoked potentials).*
>
> By the way I could not find Smith and Wave (1995)* European Journal of
> Anaesthesiology. *If you have a copy, please let me know.
>
> Makoto
> --
> Makoto Miyakoshi
> Swartz Center for Computational Neuroscience
> Institute for Neural Computation, University of California San Diego
>
> _______________________________________________
> Eeglablist page: http://sccn.ucsd.edu/eeglab/eeglabmail.html
> To unsubscribe, send an empty email to eeglablist-unsubscribe at sccn.
> ucsd.edu
> For digest mode, send an email with the subject "set digest mime" to
> eeglablist-request at sccn.ucsd.edu
>
--
Michael D. Nunez
Associate Specialist (Neuroscientist)
Laboratory of Computational and Translational Neuroscience
<http://lopour.eng.uci.edu/>
Dept. of Biomedical Engineering
Human Neuroscience Lab <http://hnl.ss.uci.edu>
Cognition and Individual Differences Lab <http://www.cidlab.com/>
Dept. of Cognitive Sciences
University of California, Irvine