[Eeglablist] Movement artifact = Liquid junction potential
Johanna Wagner
joa.wagn at gmail.com
Fri May 4 10:39:34 PDT 2018
Hi Makoto,
there are a number of recent papers that have looked at artefacts during
full body motion, some of them using also accelerometers placed at EEG
electrodes. Several of these papers use phantom heads and swim caps (see
first four papers) as to block brain EEG activity and identify pure
movement artefacts...
Symeonidou, E. R., Nordin, A. D., Hairston, W. D., & Ferris, D. P. (2018).
Effects of Cable Sway, Electrode Surface Area, and Electrode Mass on
Electroencephalography Signal Quality during Motion. *Sensors*, *18*(4),
1073.
Oliveira, A. S., Schlink, B. R., Hairston, W. D., König, P., & Ferris, D.
P. (2016). Induction and separation of motion artifacts in EEG data using a
mobile phantom head device. *Journal of neural engineering*, *13*(3),
036014.
Snyder, K. L., Kline, J. E., Huang, H. J., & Ferris, D. P. (2015).
Independent component analysis of gait-related movement artifact recorded
using EEG electrodes during treadmill walking. *Frontiers in human
neuroscience*, *9*, 639.
Kline, J. E., Huang, H. J., Snyder, K. L., & Ferris, D. P. (2015).
Isolating gait-related movement artifacts in electroencephalography during
human walking. *Journal of neural engineering*, *12*(4), 046022.
Castermans, T., Duvinage, M., Cheron, G., & Dutoit, T. (2014). About the
cortical origin of the low-delta and high-gamma rhythms observed in EEG
signals during treadmill walking. *Neuroscience letters*, *561*, 166-170.
Gwin, J. T., Gramann, K., Makeig, S., & Ferris, D. P. (2010). Removal of
movement artifact from high-density EEG recorded during walking and running.
*Journal of neurophysiology*, *103*(6), 3526-3534.
Johanna
2018-05-03 14:22 GMT-07:00 Makoto Miyakoshi <mmiyakoshi at ucsd.edu>:
> Dear Michael,
>
> Thank you for your comment.
> I'm no authority of this issue, but generally speaking, if 'Movement
> artifact = Liquid junction potential', motion artifact reflects the amount
> of displacement of the electrode against contacting body of fluid. This
> seems the movement artifact represents the first order derivative of the
> amount of the electrode displacement. Slow motion artifact could be
> generated by a slow motion, and sudden potential difference spikes could be
> generated by sudden, jerky motion.
>
> I'm curious to see if motion artifact time-series correlates with
> accelerometer measurements recorded from the electrode.
>
> Makoto
>
>
>
> On Thu, May 3, 2018 at 12:54 PM, Michael D. Nunez <mdnunez1 at uci.edu>
> wrote:
>
>> Thank you Makoto.
>>
>> In EEG recordings do we expect movement artifact (due to liquid junction
>> potential) to be expressed at slow frequencies or sudden potential
>> difference spikes (i.e. temporary electrical discontinuities)? I have found
>> empirical evidence of both.
>>
>> Example references:
>>
>> Isolating gait-related movement artifacts in electroencephalography
>> during human walking
>> <https://www.researchgate.net/profile/Kristine_Snyder/publication/278792694_Isolating_gait-related_movement_artifacts_in_electroencephalography_during_human_walking/links/55873f3e08ae71f6ba914812.pdf>
>>
>> Electroencephalography (EEG): Neurophysics, Experimental Methods, and
>> Signal Processing
>> <https://www.researchgate.net/profile/Michael_Nunez4/publication/290449135_Electroencephalography_EEG_neurophysics_experimental_methods_and_signal_processing/links/57bf32c908ae2f5eb32e82a9/Electroencephalography-EEG-neurophysics-experimental-methods-and-signal-processing.pdf>
>>
>>
>>
>> On Thu, May 3, 2018 at 12:20 PM, Makoto Miyakoshi <mmiyakoshi at ucsd.edu>
>> wrote:
>>
>>> Dear colleagues,
>>>
>>> Let me share this info.
>>>
>>> E. Huigen (2000) Noise in biopotential recording using surface electrodes
>>> https://www.semanticscholar.org/paper/Noise-in-biopotential-
>>> recording-using-surface-Huigen/8fc0837f7af0a36b19799139d9b763969057b985
>>>
>>> *%%%%%%%%%%%%%%%%%%%%%*
>>>
>>> *2.4 Motion artifact*
>>> Movement can cause changes in the potentials that are created when an
>>> electrode is applied to the skin. Normally, when the patient is relaxed,
>>> and high quality electrodes are used, the recording is not distorted by
>>> motion artifact. Brinkman (1993) has found no significant correlation
>>> between the intentional movement of the arm and the noise signal. The
>>> mechanisms that can cause motion artifact are described next.
>>>
>>> *Liquid junction potential variations by electrode movement*
>>> The various phase junctions in the electrode-electrolyte-skin interface
>>> all cause junction potentials, sensitive to motion artifact. The
>>> skin-electrolyte interface can cause artifacts of 400-600 μV when the
>>> electrode is moved parallel to the skin surface (Smith and Wace, 1995).
>>> When the electrode is moved perpendicular to the skin the potential changes
>>> can be up to 900 μV. Firm attachment to the skin can reduce the
>>> potential changes. The electrode-electrolyte interface also produces
>>> artifacts when mechanically disturbed. Gatzke (1974, as described in
>>> Webster, 1984) measured a 15 mV potential change when a pure silver
>>> electrode is moved in electrolyte. Coating with silver chloride, thus
>>> creating a stable double layer, produces a 10-fold reduction of the
>>> artifact. Further reduction (up to negligible value) can be achieved by
>>> recessing the electrode-electrolyte interface in a protective cup, in which
>>> a sponge soaked in gel is placed (figure 2-2).
>>>
>>> *Skin potential changes*
>>> Earlier, the stratum corneum and the barrier layer have been identified
>>> as the major sources of the impedance of the skin. Webster (1984) has also
>>> observed a potential difference between the inside and the outside of the
>>> barrier layer. Van Wijk van Brievingh (1988) however states that this
>>> skin potential is a liquid junction potential between deeper skin layers
>>> and the electrolyte. The skin potential has a typical value of 30 mV at
>>> the thorax. Stretching of the skin decreases the skin potential to about 25
>>> mV. This artifact can be reduced to negligible value by abrading the skin.
>>> Webster gives 20 strokes with fine sandpaper as an indication. Shackel
>>> (1959, as described in Geddes and Baker, 1989) developed a method for
>>> shorting the skin potential, called the skin-drilling technique. The
>>> epidermis is eroded using a dental burr. The capillaries remain
>>> undisturbed, so no blood is drawn. Unfortunately, the epidermis is also the
>>> layer that protects the skin from irritation. A mild electrode gel has to
>>> be used to prevent
>>> unwanted effects.
>>>
>>> %%%%%%%%%%%%%%%%%%%%%
>>>
>>> Wikipedia 'liquid junction potential'
>>> https://en.wikipedia.org/wiki/Liquid_junction_potential
>>>
>>> In Huigen, Peper, Grimbergen (2002) Med. biol. Eng, they described:
>>>
>>> *RECORDINGS OF biomedical signals from the body surface often
>>> contain a substantial noise component. This noise signal can
>>> severely impair the resolution of biomedical recordings as its
>>> magnitude can be as high as 10-60 uVp_p (GONDRAN et aL, 1996).
>>> This is in the range of EEG potentials and is at least ten
>>> times as high as several types of evoked potentials (e.g. visual
>>> evoked potentials or somatosensory evoked potentials).*
>>>
>>> By the way I could not find Smith and Wave (1995)* European Journal of
>>> Anaesthesiology. *If you have a copy, please let me know.
>>>
>>> Makoto
>>> --
>>> Makoto Miyakoshi
>>> Swartz Center for Computational Neuroscience
>>> Institute for Neural Computation, University of California San Diego
>>>
>>> _______________________________________________
>>> Eeglablist page: http://sccn.ucsd.edu/eeglab/eeglabmail.html
>>> To unsubscribe, send an empty email to eeglablist-unsubscribe at sccn.uc
>>> sd.edu
>>> For digest mode, send an email with the subject "set digest mime" to
>>> eeglablist-request at sccn.ucsd.edu
>>>
>>
>>
>>
>> --
>> Michael D. Nunez
>> Associate Specialist (Neuroscientist)
>> Laboratory of Computational and Translational Neuroscience
>> <http://lopour.eng.uci.edu/>
>> Dept. of Biomedical Engineering
>> Human Neuroscience Lab <http://hnl.ss.uci.edu>
>> Cognition and Individual Differences Lab <http://www.cidlab.com/>
>> Dept. of Cognitive Sciences
>> University of California, Irvine
>>
>
>
>
> --
> Makoto Miyakoshi
> Swartz Center for Computational Neuroscience
> Institute for Neural Computation, University of California San Diego
>
> _______________________________________________
> Eeglablist page: http://sccn.ucsd.edu/eeglab/eeglabmail.html
> To unsubscribe, send an empty email to eeglablist-unsubscribe at sccn.
> ucsd.edu
> For digest mode, send an email with the subject "set digest mime" to
> eeglablist-request at sccn.ucsd.edu
>
--
--
Johanna Wagner, PhD
Postdoctoral Researcher
Swartz Center for Computational Neuroscience
Institute for Neural Computation
University of California San Diego
http://scholar.google.at/citations?user=vSJYGtcAAAAJ&hl=en
<http://sccn.ucsd.edu/>