[Eeglablist] How many electrodes tolerated for interpolation?
Richards, John
RICHARDS at mailbox.sc.edu
Tue Nov 18 18:53:42 PST 2025
1--source level EEG analysis is a transformation of the scalp potential to source locations. If the scalp potential has functionally significant relations with experimental variables or clinical conditions, so will the source locations. Probably the same if the scalp potential is NS for experimental variables or clinical conditions, so will the source level be NS. Reliability might be different, since each source location data is the sum of several electrodes. Might be more reliable.
2--I'm not totally sure what you want to do with interpolation. There are MANY useful reasons for using it. Typically EEG is spread across a number of electrodes, and not in just one location. If you have a bad channel or channels, you can safely estimate an amplitude in that channel by appropriate interpolation from groups of electrodes (e.g., spherical spline). Since adjacent groups of electrodes usually have a similar signal, the interpolation of data for missing channels is appropriate.
3--Many researchers studying infant subjects or special populations (neurodevelopmental disorders) cannot get a large enough set of data for ERP, or other, and want to optimize their subjects' data. So they use some kind of rubric for estimating data from missing channels, e.g., no more than 10 channels of a 128 channel system with missing data.
4--The "lead-field" matrix in source analysis is computed from conductivities, electrodes, and head media (FEM, BEM). Its sometimes onerous to compute. E.g., all 128 electrodes are used to compute the LF and inverse spatial filter. You cannot use the LF/inverse spatial filter with < the electrode positions used to compute it. So using the same rubric (missing channels <= 10) you can use the data in the source computation when there are missing channels.
5--Nearly all people using a non-10-10 configuration (like EGI) wish to present data in the 10-10 nomenclature. Typically one or more electrodes "near" the 10-10 electrode are used. I have been using spherical spline interpolation from the 128 channels to the 81 10-10 channels, using the 128 channel location and the 10-10 channel location to make the calculation. I and colleagues have used this successfully in a number of papers.
John
***********************************************
John E. Richards
Carolina Distinguished Professor
Department of Psychology
University of South Carolina
Columbia, SC 29208
Dept Phone: 803 777 2079
Fax: 803 777 9558
Email: richards-john at sc.edu
https://urldefense.com/v3/__https://jerlab.sc.edu/__;!!Mih3wA!GB28o9-Jv8lJb3cT872SQ6Uta26iWuuI0NUIiUKA4ssRKyEXpBQ1U4t7XqnfPAcDqV4H5WhvVPXifO0CI4IJtIAI-tXACQ$
***********************************************
-----Original Message-----
From: eeglablist <eeglablist-bounces at sccn.ucsd.edu> On Behalf Of 장진원 via eeglablist
Sent: Tuesday, November 18, 2025 2:40 PM
To: smakeig at gmail.com
Cc: eeglablist at sccn.ucsd.edu; fernandez luis <isabelyluis2007 at me.com>
Subject: Re: [Eeglablist] How many electrodes tolerated for interpolation?
Dear Prof. Makeig,
I appreciate your suggestion. Is source-level EEG analysis more reliable than scalp-level analysis? I wonder whether source-level EEG analysis (such as dipole fitting) achieves high test-retest reliability and replicability over different datasets, especially in clinical populations (depression vs healthy controls, as an example)? I am not sure whether source-level analysis is applicable in clinical research.
Best Regards,
Jinwon Chang
2025년 11월 18일 (화) 오후 1:19, Scott Makeig <smakeig at gmail.com>님이 작성:
> I would question the value of channel interpolation - except for
> making smooth pictures of particular scalp distributions. Individual
> electrode-signal-difference scalp channels are inherently vague
> measures that sum electrical activities generated in many unrelated parts of cortex.
> Scalp channel signals are not more deserving of attention than
> individual radio-frequency channels in an fMRI system ...
>
> I do understand that (at least in the US) clinicians need to work in
> terms that insurance companies will reimburse. But cortical
> source-resolved EEG recording and analysis is technically quite
> feasible, as our work at SCCN over the last 30 years has abundantly
> demonstrated -- and allows more exact interpretation (and high
> statistical certainty) than scalp-level data interpretation.
>
> Scott Makeig
>
> On Tue, Nov 18, 2025 at 9:41 AM 장진원 via eeglablist <
> eeglablist at sccn.ucsd.edu> wrote:
>
>> Thank for your kind reply. I understand it.
>>
>> Best Regards,
>> Jinwon
>>
>> On Mon, Nov 17, 2025 at 7:23 PM fernandez luis via eeglablist <
>> eeglablist at sccn.ucsd.edu> wrote:
>>
>> > > Hi Jinwon,
>> > >
>> > > Accurate EEG channel interpolation is methodologically feasible
>> across a
>> > broad spectrum of montage densities, including low-density
>> configurations
>> > such as 19- and 32-channel systems, intermediate-density arrays
>> > such as
>> 64
>> > channels, and high-density systems such as 128 channels.
>> > Importantly,
>> the
>> > methodological validity of interpolation does not primarily depend
>> > on
>> the
>> > absolute number of electrodes, but on the availability of
>> > non-artifacted neighboring electrodes with adequate spatial
>> > distribution surrounding
>> the
>> > channel to be reconstructed.
>> > >
>> > >
>> > > EEG interpolation is a spatial estimation procedure in which the
>> signal
>> > of an artifacted electrode is reconstructed using mathematically
>> weighted
>> > contributions from the nearest clean surrounding electrodes.
>> Accordingly,
>> > if the spatially adjacent electrodes are also artifacted,
>> > interpolation becomes unreliable or methodologically inappropriate,
>> > because the reconstruction process would be driven by distorted
>> > input data—violating key assumptions underlying spherical spline
>> > interpolation,
>> inverse-distance
>> > weighting, and other spatial estimation algorithms.
>> > >
>> > >
>> > > In low-density montages (e.g., 19 or 32 channels), interpolation
>> remains
>> > technically feasible; however, the reduced spatial sampling
>> > inherently limits the anatomical precision and spatial granularity
>> > of the reconstructed signal. Nevertheless, interpolation in these
>> > systems can yield clinically acceptable results as long as the
>> > electrodes used as sources for reconstruction are clean, stable,
>> > and sufficiently
>> distributed
>> > around the artifacted location.
>> > >
>> > >
>> > > Intermediate-density systems such as 64-channel EEG offer
>> > > improved
>> > spatial resolution that allows more accurate reconstruction of
>> > missing channels due to enhanced scalp coverage. High-density
>> > montages, particularly 128-channel EEG systems, provide dense and
>> > homogeneous
>> spatial
>> > sampling, minimizing interpolation error and generating
>> > reconstructions that are more physiologically plausible and quantitatively reliable.
>> This
>> > level of spatial resolution is advantageous for applications
>> > requiring high-fidelity scalp mapping, microstate analysis,
>> > connectivity
>> estimation,
>> > and source localization.
>> > >
>> > >
>> > > Despite differences in resolution across montage densities, a
>> > fundamental methodological requirement remains invariant:
>> > interpolation must be performed exclusively using clean,
>> > non-artifacted surrounding electrodes. Reconstruction based on
>> > artifacted neighbors compromises the physiological validity of the
>> > estimated signal and undermines the mathematical assumptions intrinsic to spatial interpolation algorithms.
>> > >
>> > >
>> > > Technical Comparison: Conventional vs. High-Density EEG
>> > > Interpolation
>> > >
>> > >
>> > > 1. Interpolation in conventional EEG (19–32 channels)
>> > >
>> > >
>> > > Low-density EEG systems rely on sparse spatial sampling, which
>> > > imposes
>> > several methodological constraints:
>> > >
>> > >
>> > > • Wide inter-electrode spacing
>> > >
>> > > • Higher vulnerability to local contamination
>> > >
>> > > • Reduced capacity to capture rapid spatial changes
>> > >
>> > > • Acceptable but limited reliability
>> > >
>> > >
>> > > 2. Interpolation in high-density EEG (64–128 channels)
>> > >
>> > >
>> > > High-density EEG (HD-EEG) significantly enhances the reliability
>> > > of
>> > interpolation due to:
>> > >
>> > >
>> > > • Dense and homogeneous spatial sampling
>> > >
>> > > • Robustness to isolated corrupted channels
>> > >
>> > > • Improved modeling of spatial gradients
>> > >
>> > > • Near-physiological reconstruction in 128-channel systems
>> > >
>> > > Best
>> > > Luis Fernandez, MSc
>> > > Clinical Neuropsychologist
>> > >
>> >
>> > > El 17 nov 2025, a las 20:41, 장진원 via eeglablist <
>> > eeglablist at sccn.ucsd.edu> escribió:
>> > >
>> > > Hi all,
>> > >
>> > > I'm a clinical psychiatrist, so I am not really familiar with
>> engineering
>> > > concept of interpolation. I believe in high-density setting (128
>> channel)
>> > > interpolation of a few channels are acceptable, but what if more
>> > > than
>> 10
>> > > bad channels in 64 channel setting? Is it tolerable or
>> > > detrimental for maintenance of true signal?
>> > >
>> > > Best regards,
>> > > Jinwon Chang
>> > > _______________________________________________
>> > > To unsubscribe, send an empty email to
>> > eeglablist-unsubscribe at sccn.ucsd.edu or visit
>> > https://sccn.ucsd.edu/mailman/listinfo/eeglablist .
>> >
>> > _______________________________________________
>> > To unsubscribe, send an empty email to
>> > eeglablist-unsubscribe at sccn.ucsd.edu or visit
>> > https://sccn.ucsd.edu/mailman/listinfo/eeglablist .
>> _______________________________________________
>> To unsubscribe, send an empty email to
>> eeglablist-unsubscribe at sccn.ucsd.edu or visit
>> https://sccn.ucsd.edu/mailman/listinfo/eeglablist .
>
>
>
> --
> Scott Makeig, Research Scientist and Director, Swartz Center for
> Computational Neuroscience, Institute for Neural Computation,
> University of California San Diego, La Jolla CA 92093-0559,
> https://urldefense.com/v3/__http://sccn/__;!!Mih3wA!GB28o9-Jv8lJb3cT872SQ6Uta26iWuuI0NUIiUKA4ssRKyEXpBQ1U4t7XqnfPAcDqV4H5WhvVPXifO0CI4IJtIAIWh7cwg$ .
> ucsd.edu%2F~scott&data=05%7C02%7CRICHARDS%40mailbox.sc.edu%7C5cb99077f
> 3e94073b45708de26f9ce0e%7C4b2a4b19d135420e8bb2b1cd238998cc%7C0%7C0%7C6
> 38991051525194165%7CUnknown%7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYi
> OiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ%3D%3D%7C0%7
> C%7C%7C&sdata=VMvxPmleLR8%2BlLATkWe0TD5%2Fd9pXsCnaj24H6eslbd0%3D&reser
> ved=0
>
_______________________________________________
To unsubscribe, send an empty email to eeglablist-unsubscribe at sccn.ucsd.edu or visit https://sccn.ucsd.edu/mailman/listinfo/eeglablist .
More information about the eeglablist
mailing list