Hi Kathryn,

On linked mastoids, I agree with Makoto. Based on work on alpha asymmetries, it is strongly recommended not to use linked mastoid referencing for two related but distinct reasons:

First, linked mastoids suffer from alpha “mirroring”: occipital alpha dipoles produce activity of opposite polarity at sites 180 degrees apart, and since spectral power summarizes oscillation magnitude without regard to polarity, occipital alpha bleeds into frontal leads through the reference. This contaminates frontal alpha asymmetry measures with signal unrelated to frontal sources (Smith, E. E., Reznik, S. J., Stewart, J. L., & Allen, J. J. B. (2017). Assessing and conceptualizing frontal EEG asymmetry: An updated primer on recording, processing, analyzing, and interpreting frontal alpha asymmetry).

Second, if a lateralized neural source drives the two mastoids in opposite polarity, linking them creates a reference that is itself contaminated by the signal of interest, which can suppress or distort asymmetry scores (Hagemann, D., Naumann, E., & Thayer, J. F. (2001). The quest for the EEG reference revisited: A glance from brain asymmetry research).

The CSD (current source density) transformation avoids both issues by computing the second spatial derivative of voltage, attenuating distal volume-conducted sources and providing a reference-free estimate. Although CSD also has downsides, which were discussed in a recent eeglablist thread. I can try to find it if you want to learn more.

You can perform CSD in one line in eeglab with this code: https://urldefense.com/v3/__https://github.com/amisepa/csd_transfrom__;!!Mih3wA!D-IJVG1iQfvM3mGG1a0VAvcyK6Xydhoi-GPhWN60iDa7XwfiAWK7YvhCoTrhdKrE-kTPfcJ7V2ot9FMsWN8rw9he4g$ 

Or with Rest/infinity method here:
https://urldefense.com/v3/__https://github.com/amisepa/reference_infinity__;!!Mih3wA!D-IJVG1iQfvM3mGG1a0VAvcyK6Xydhoi-GPhWN60iDa7XwfiAWK7YvhCoTrhdKrE-kTPfcJ7V2ot9FMsWN8XO01aNw$ 
(Also available in eeglab’s extension manager, search ref_rest_cmd)

Cedric

Sent from Proton Mail for iOS.

-------- Original Message --------
On Monday, 05/04/26 at 07:25 Kathryn Bolton via eeglablist <eeglablist at sccn.ucsd.edu> wrote:
Hello Makoto,

Thank you so very much for your advice and sharing those resources with me!
I have a follow-up question regarding referencing to the mastoids: in my
script I re-reference to the mastoids, but then remove them from before the
ICA (script pasted below). I do all this before the high-pass filter and
downsampling. I'm using a 64-channel BioSemi, so I could try your
suggestion of using Dezhong Yao's REST, but what might that look like in my
script?
% 3. Re-reference to channels 65 and 66
EEG = pop_reref(EEG, [65 66]);
% 4. Keep only scalp channels 1 through 64
% This removes the reference channels and any other extra sensors
EEG = pop_select(EEG, 'channel', 1:64);
% Save channel locations for later interpolation
original_chanlocs = EEG.chanlocs;
save([filename '_chanlocs.mat'], 'original_chanlocs');
With many thanks, Kathryn

*Kathryn Bolton, MA*

Pronouns: she/her

PhD Candidate | Clinical Psychology

Department of Psychology | Cognitive Aging Lab (CAL)

Toronto Metropolitan University

*E-mail:* *kathryn.bolton at t <kathryn.bolton at ryerson.ca>orontomu.ca
<https://urldefense.com/v3/__http://orontomu.ca__;!!Mih3wA!DnWNR0lZhJPKbmyrmmnVW0Mrle4NIMGZz_fnSZy0zMvAGAa6gL1PNlVkpkc7dOe56-66IdfF1UYesT1-7xpPXavznSFMjXOqu64$ >*

*Website*: *https://urldefense.com/v3/__https://psychlabs.torontomu.ca/cal/portfolio/kathryn-bolton/__;!!Mih3wA!DnWNR0lZhJPKbmyrmmnVW0Mrle4NIMGZz_fnSZy0zMvAGAa6gL1PNlVkpkc7dOe56-66IdfF1UYesT1-7xpPXavznSFM_xtYgn0$
<https://urldefense.com/v3/__https://psychlabs.torontomu.ca/cal/portfolio/kathryn-bolton/__;!!Mih3wA!DnWNR0lZhJPKbmyrmmnVW0Mrle4NIMGZz_fnSZy0zMvAGAa6gL1PNlVkpkc7dOe56-66IdfF1UYesT1-7xpPXavznSFM_xtYgn0$ >*


*Due to the volume of emails I receive, I will do my best to respond within
2 business days. If you require an immediate response or if it is an
emergency, please put "urgent" in the subject line. Thank you for your
understanding. *


On Wed, Apr 29, 2026 at 8:01 PM Makoto Miyakoshi via eeglablist <
eeglablist at sccn.ucsd.edu> wrote:

> Hi Kathryn,
>
> But my question about filtering is whether you recommend applying only a
> high-pass
> filter prior to ICA (as above), and then applying the full bandpass (e.g.,
> including a low-pass) later, prior to PSD analysis?
>
>
> Andreas Widmann answered that question once and for all in this 2015 paper.
> 10.1016/j.jneumeth.2014.08.002
>
> The answer is, apply high-pass and low-pass separately, with an
> (inevitably) short transition band width (TBW) to high-pass filter, and a
> gentle TBW for low-pass filter.
> For example, if you want to set a band-pass filter 0.5-55 Hz @-6dB, then
>
> Highpass: cutoff 0.5 Hz (TBW 0.5 * 2 = 1.0)
> Lowpass: cutoff 55 Hz (TBW 5, which makes pass-band edge at 55-5/2 = 52.5
> Hz, and maximum suppression starts at 55+5/2 = 57.5 Hz)
>
> For confirming definitions of terminologies, see this page. This page is a
> summary of my questions and his answers.
>
> https://urldefense.com/v3/__https://eeglab.org/others/Firfilt_FAQ.html__;!!Mih3wA!HPZkz51i3HC_zP7sK-ZAnbc3e7bwYY3dwxbgLZn8v25vAVwmBxhdk6YfF87fgdUWoKXpUNspBYhMnKmFvicAeN5kvEY$
>
> The reason why you want to do this is that you do not want to use an
> unnecessarily steep slope in the low-pass end, which makes the filter order
> unnecessarily long.
>
> Re-reference to linked mastoids prior to ICA (BioSemi recommendations)
>
> My advice: do not use linked mastoids. Physically linked mastoids/earlobes
> is out of the question, this is the worst because electrically your
> subject's head is no longer shaped like a human head (heads are connected
> bilaterally)!!! Digitally linked mastoids/earlobes is also problematic for
> the following reason (from *Electric Fields of the Brain* p. 294, Nunez and
> Srinivasan 2006).
>
> The mathematical linked-ears or linked-mastoids reference may seem to
> provide a "solution" to these problems associated with the physical linked
> reference. However, does the original goal of using the average potential
> of the two ears as the reference make sense? The motivation for this
> procedure is not based on any physical properties of head volume
> conduction. The measured potential difference between a pair of electrodes
> depends on sources located near both electrode positions. With the
> mathematical linked-ears reference, the recorded potential depends on
> sources at three different locations. This approach may further complicate
> the interpretation of scalp potentials and possible source locations rather
> than simplifying it. One rationalization often argued to support the
> linked-ears or linked-mastoids reference is its purported tendency to be a
> "symmetric" reference with respect to both brain hemispheres, thereby
> providing a tool for characterizing hemispheric asymmetries in EEG studies.
> However, this is not generally true; the effect of the average mastoids
> reference is artificially to correlate data from recording electrodes near
> the two mastoids (Srinivasan et al. 1998), potentially reducing estimates
> of dynamic hemispheric source asymmetries. We will consider several
> simulations in the next section where we introduce the average reference
> and compare it to the mathematical linked-ears reference.
>
>
> Instead, use either average reference (if you have > 64 ch) or Dezhong
> Yao's REST (if you use the low number of electrodes, you might want to use
> this, but it is conceptually much more complicated than average reference).
>
> Makoto
>
> On Tue, Apr 28, 2026 at 11:35 PM Kathryn Bolton via eeglablist <
> eeglablist at sccn.ucsd.edu> wrote:
>
> > Hi EEGLAB mailing list,
> >
> > I hope you're having a good week so far! I’m a PhD student working on an
> > EEG study examining affective responses to music in older adults.
> > Participants listen to multiple 1-minute music samples, and I plan to
> > compute frontal alpha asymmetry (FAA) and beta/alpha ratio (PSD-based
> > measures) and relate these to valence/arousal ratings.
> >
> > I’ve structured my pipeline into three stages, partly to distribute work
> > across research assistants:
> >
> >    - *Script 1:* preprocessing + ICA (run by RAs; outputs ICA-ready
> >    datasets)
> >    - *Script 2:* manual ICA inspection/cleaning (topography, spectrum,
> >    timecourse), followed by channel interpolation
> >    - *Script 3:* epoching, PSD computation, and extraction of FAA and
> >    beta/alpha ratio
> >
> > In my original pipeline, I noticed that the component maps used for
> > inspection in Script 2 appeared a bit off (e.g., nearly uniform colour
> > across components). I’ve since revised Script 1 to the following
> workflow:
> >
> >    - Re-reference to linked mastoids prior to ICA (BioSemi
> recommendations)
> >    - High-pass filter at 0.5 Hz only, removing a 50 Hz low-pass filter at
> >    this stage
> >    - Downsample to 512 Hz
> >    - Automated detection and removal of bad channels (logged), prior to
> ICA
> >    - Run ICA using rank-adjusted PCA
> >
> > This seems to have resolved the issue with the component maps. But my
> > question about filtering is whether you recommend applying only a
> high-pass
> > filter prior to ICA (as above), and then applying the full bandpass
> (e.g.,
> > including a low-pass) later, prior to PSD analysis?
> >
> > I want to ensure that the filtering approach is appropriate both for ICA
> > decomposition and for downstream frequency-based analyses (FAA and
> > beta/alpha ratio).
> >
> > Thank you very much for your time and advice, I really appreciate it!
> >
> > Best regards,
> > Kathryn
> >
> >
> > *Kathryn Bolton, MA*
> >
> > Pronouns: she/her
> >
> > PhD Candidate | Clinical Psychology
> >
> > Department of Psychology | Cognitive Aging Lab (CAL)
> >
> > Toronto Metropolitan University
> >
> > *E-mail:* *kathryn.bolton at t <kathryn.bolton at ryerson.ca>orontomu.ca
> > <
> >
> https://urldefense.com/v3/__http://orontomu.ca__;!!Mih3wA!HEf3RTkd7gvWzdftBOsrCvgP_ahawX4XGp6D3OMOOK2ck_zqZyrcOpLjBJdpl7-leggZto82bqKspbCJ49hb7OlEv45uNvcNR5M$
> > >*
> >
> > *Website*: *
> >
> https://urldefense.com/v3/__https://psychlabs.torontomu.ca/cal/portfolio/kathryn-bolton/__;!!Mih3wA!HEf3RTkd7gvWzdftBOsrCvgP_ahawX4XGp6D3OMOOK2ck_zqZyrcOpLjBJdpl7-leggZto82bqKspbCJ49hb7OlEv45uwX-37m8$
> > <
> >
> https://urldefense.com/v3/__https://psychlabs.torontomu.ca/cal/portfolio/kathryn-bolton/__;!!Mih3wA!HEf3RTkd7gvWzdftBOsrCvgP_ahawX4XGp6D3OMOOK2ck_zqZyrcOpLjBJdpl7-leggZto82bqKspbCJ49hb7OlEv45uwX-37m8$
> > >*
> >
> >
> > *Due to the volume of emails I receive, I will do my best to respond
> within
> > 2 business days. If you require an immediate response or if it is an
> > emergency, please put "urgent" in the subject line. Thank you for your
> > understanding. *
> > _______________________________________________
> > To unsubscribe, send an empty email to
> > eeglablist-unsubscribe at sccn.ucsd.edu or visit
> > https://sccn.ucsd.edu/mailman/listinfo/eeglablist   .
> _______________________________________________
> To unsubscribe, send an empty email to