[Eeglablist] Persistent 20 Hz Harmonic Line Noise in EEG Data

Maeghal Jain maeghaljain at gmail.com
Sun May 17 22:24:38 PDT 2026


Hi Eric,

Thank you very much for the detailed and helpful reply.

Over the weekend I sat and wrote some version of this for my dataset (
https://urldefense.com/v3/__https://github.com/csea-lab/csea-lab/blob/refs*2Fheads*2Fmaster/Libas*2FLB3*2FLB3_prepro_eegfmri_1.m__;JSUlJQ!!Mih3wA!AecfBxaMRbnkXpSJ4rvweP9891s7UjaDa-pJhM1RcUuiQnhGpaID0zFJ0jPVfcWaL8UfyBcfEsF2hxpXeaFh6i6OJQ$ )
and now the PSD and topomap both look a bit better.

PSD: https://urldefense.com/v3/__https://ibb.co/kg9h9s8P__;!!Mih3wA!AecfBxaMRbnkXpSJ4rvweP9891s7UjaDa-pJhM1RcUuiQnhGpaID0zFJ0jPVfcWaL8UfyBcfEsF2hxpXeaGVkjJ2IQ$ 
Topomap: https://urldefense.com/v3/__https://ibb.co/WvdW3xFB__;!!Mih3wA!AecfBxaMRbnkXpSJ4rvweP9891s7UjaDa-pJhM1RcUuiQnhGpaID0zFJ0jPVfcWaL8UfyBcfEsF2hxpXeaGEiH6G3g$ 

On average template subtraction: I assume this is not\ the same as average
re-referencing (which I tried in multiple implementations with no
improvement on the 20 Hz harmonic). Could you clarify or point me to the
exact code/steps for average (or moving-window) template subtraction? I’d
be very grateful for any example code snippet.

Recommended papers: Could you list one or two papers you find most relevant
or useful for this kind of preprocessing?

I would genuinely appreciate the opportunity to discuss this further. I’ll
send you a direct message/calendar invite shortly to book a short call at a
convenient time for you.

Thank you again for your time and expertise!

Best regards,
Maeghal Jain
BISE Lab

On Fri, May 15, 2026, 1:33 AM Rawls, Eric <rawlse at uncw.edu> wrote:

> Hi Maeghal,
>
> That changes things substantially. There are very strong noise sources in
> simultaneous EEG-fMRI that cannot be dealt with using standard methods.
>
> First and most important - have you processed the data to remove gradient
> artifacts? There are a number of ways to this, from average template
> subtraction (simplest), to moving window template subtraction, to
> augmentation with principal components of the template (this is implemented
> in the EEGLAB plugin fMRIB2.1 which I have found high performance with).
> Note that ICA is entirely unsuitable for gradient artifact removal.
>
> Second, you will find your data contaminated with ballistocardiographic
> artifact, although that is not likely to cause your 20Hz contamination. You
> will need to detect heartbeats from a (hopefully dedicated) ECG channel to
> remove these artifacts which mostly occur at low frequencies and are
> magnitudes larger than EEG. This is also implemented in fMRIB2.1 for EEGLAB.
>
> Finally, note that you will encounter additional noise from the helium
> pump and other sources. These sources of noise are not easy to remove from
> the EEG, and while the pump can be turned off, we found that this caused
> instabilities in our magnet's temperature so instead we focused analysis on
> frequencies outside of the helium pump's native frequency.
>
> On a side note, I have worked with simultaneous EEG-fMRI data for several
> years now, and would be happy to discuss artifact removal strategies and
> aoftware solutions for these methods. Please let me know.
>
> Best, E
>
>
> Eric Rawls, Ph.D. (he/him/his)
>
> Assistant Professor | Department of Psychology
> Director | Brain, Data, and Causality (BDC) Lab
> College of Science and Engineering
> University of North Carolina Wilmington
>
> Adjunct Assistant Professor | Department of Psychiatry and Behavioral
> Sciences
> College of Medicine
> University of Minnesota Twin Cities
>
> The University of North Carolina Wilmington rests on historic lands of the
> Cape Fear Indians and the Waccamaw Siouan people.
>
> The University of Minnesota rests on historic lands of the Dakota people.
>
> NOTICE: Emails sent and received in the course of university business are
> subject to the North Carolina Public Records Act (N.C.G.S. §132-1 et seq.)
> and may be released to the public unless an exception applies.
>
>
> <https://urldefense.com/v3/__https://outlook.office.com/bookwithme/user/82c92b80a22d45b0a310ddc81d7cf83c@uncw.edu?anonymous&ep=signature&ismsaljsauthenabled__;!!Mih3wA!AecfBxaMRbnkXpSJ4rvweP9891s7UjaDa-pJhM1RcUuiQnhGpaID0zFJ0jPVfcWaL8UfyBcfEsF2hxpXeaHD5bVZjA$ >
> Book time to meet with me
> <https://urldefense.com/v3/__https://outlook.office.com/bookwithme/user/82c92b80a22d45b0a310ddc81d7cf83c@uncw.edu?anonymous&ep=signature&ismsaljsauthenabled__;!!Mih3wA!AecfBxaMRbnkXpSJ4rvweP9891s7UjaDa-pJhM1RcUuiQnhGpaID0zFJ0jPVfcWaL8UfyBcfEsF2hxpXeaHD5bVZjA$ >
> ------------------------------
> *From:* eeglablist <eeglablist-bounces at sccn.ucsd.edu> on behalf of
> Maeghal Jain via eeglablist <eeglablist at sccn.ucsd.edu>
> *Sent:* Thursday, May 14, 2026 10:36 AM
> *To:* Cedric Cannard <ccannard at protonmail.com>
> *Cc:* EEGLAB List <eeglablist at sccn.ucsd.edu>
> *Subject:* Re: [Eeglablist] Persistent 20 Hz Harmonic Line Noise in EEG
> Data
>
> [This email originated from outside of UNCW]
>
> Hi Cedric,
>
> Yes, this data is EEG-fMRI recorded simultaneously. That's probably why
> this must be happening.
>
> Should I remove it with ICA or I tried Zapline as well. ICA makes sense but
> am just not sure how to pick out this one component that causes it all.
>
> Thank you for raising this point.
>
> Maeghal Jain
>
> On Thu, May 14, 2026 at 9:22 AM Cedric Cannard via eeglablist <
> eeglablist at sccn.ucsd.edu> wrote:
>
> > I would follow all these great suggestions.
> > One thing that hasn’t been asked unless I missed it. Is your system
> > recording other modalities simultaneously (GSR, PPG, fNIRS, etc.)?
> >
> > If so, many PPG modules use PWM-driven LEDs pulsing at or near 20 Hz, and
> > electrical crosstalk through shared grounding or circuitry can inject
> this
> > directly into the EEG signal.
> >
> > GSR/EDA devices typically use low-frequency AC excitation, so
> > high-frequency interference is less common, but I’ve seen it with the
> > Biosemi system at 16 Hz.
> >
> > fNIRS modulates its light sources at specific frequencies (sometimes
> > 20-100 Hz) for lock-in detection. If the driving electronics share a
> power
> > supply or chassis with the EEG amplifier, that modulation frequency can
> > appear as a narrow spectral peak.
> >
> > If that’s the case, the cleanest diagnostic is to record EEG with each
> > co-sensor powered on vs. off and check whether the artifact peak appears
> > only when a given sensor is active.​​​​​​​​​​​​​​​​
> >
> >
> > Cedric
> >
> > Sent from Proton Mail for iOS.
> >
> > -------- Original Message --------
> > On Tuesday, 05/12/26 at 17:06 Rawls, Eric via eeglablist <
> > eeglablist at sccn.ucsd.edu> wrote:
> > Is the 20Hz interference visible in the power spectrum of the raw data or
> > does it only show up after some preprocessing? I have seen something like
> > this occur with specific downsampling/filtering orders in older versions
> of
> > EEGLAB but I do not recall the exact order-of-operations required to
> > recreate it.
> >
> > Best, E
> >
> >
> > Eric Rawls, Ph.D. (he/him/his)
> >
> > Assistant Professor | Department of Psychology
> > Director | Brain, Data, and Causality (BDC) Lab
> > College of Science and Engineering
> > University of North Carolina Wilmington
> >
> > Adjunct Assistant Professor | Department of Psychiatry and Behavioral
> > Sciences
> > College of Medicine
> > University of Minnesota Twin Cities
> >
> > Land Acknowledgment: I acknowledge that UNCW is located on the ancestral
> > and traditional homelands of Indigenous peoples, including the Waccamaw
> > Siouan and other tribes with enduring ties to this region such as the
> > Lumbee, Coharie, Haliwa-Saponi, and Cherokee. I honor the ongoing
> presence,
> > history, and stewardship of Native communities and offer this
> > acknowledgment in a spirit of respect and relationship.
> >
> > NOTICE: Emails sent and received in the course of university business are
> > subject to the North Carolina Public Records Act (N.C.G.S. §132-1 et
> seq.)
> > and may be released to the public unless an exception applies.
> >
> > [cid:ca52e587-5639-4806-9792-0befc92fb683]<
> >
> https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Furldefense.com*2Fv3*2F__https*3A*2F*2Foutlook.office.com*2Fbookwithme*2Fuser*2F82c92b80a22d45b0a310ddc81d7cf83c*40uncw.edu*3Fanonymous*26ep*3Dsignature*26ismsaljsauthenabled__*3B!!Mih3wA!AIhiGfU3uPoKEyYZGULe3B4YLszx9mBhQGd07e2ZK9SnCFUO7t1H_VXRrWWbSCO6Sxru1LMboIkoQD5astO1_g*24&data=05*7C02*7Crawlse*40uncw.edu*7Ce475fe096e8844b757b208deb1cc3912*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639143687887823195*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=KZuWVsIUM57R2nZ8P4ZSQXpDf0VhKHK1iuL4V9JVVso*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSU!!Mih3wA!AecfBxaMRbnkXpSJ4rvweP9891s7UjaDa-pJhM1RcUuiQnhGpaID0zFJ0jPVfcWaL8UfyBcfEsF2hxpXeaEiyTcUBg$ 
> <https://urldefense.com/v3/__https://outlook.office.com/bookwithme/user/82c92b80a22d45b0a310ddc81d7cf83c@uncw.edu?anonymous&ep=signature&ismsaljsauthenabled__;!!Mih3wA!AIhiGfU3uPoKEyYZGULe3B4YLszx9mBhQGd07e2ZK9SnCFUO7t1H_VXRrWWbSCO6Sxru1LMboIkoQD5astO1_g$>
> > >
> > Book time to meet with me<
> >
> https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Furldefense.com*2Fv3*2F__https*3A*2F*2Foutlook.office.com*2Fbookwithme*2Fuser*2F82c92b80a22d45b0a310ddc81d7cf83c*40uncw.edu*3Fanonymous*26ep*3Dsignature*26ismsaljsauthenabled__*3B!!Mih3wA!AIhiGfU3uPoKEyYZGULe3B4YLszx9mBhQGd07e2ZK9SnCFUO7t1H_VXRrWWbSCO6Sxru1LMboIkoQD5astO1_g*24&data=05*7C02*7Crawlse*40uncw.edu*7Ce475fe096e8844b757b208deb1cc3912*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639143687887867416*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=12ZjqlDVKxV4uaMoXWhzpJp*2BVHAtp2eWaclIfulYMQQ*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUl!!Mih3wA!AecfBxaMRbnkXpSJ4rvweP9891s7UjaDa-pJhM1RcUuiQnhGpaID0zFJ0jPVfcWaL8UfyBcfEsF2hxpXeaF8a3aH_Q$ 
> <https://urldefense.com/v3/__https://outlook.office.com/bookwithme/user/82c92b80a22d45b0a310ddc81d7cf83c@uncw.edu?anonymous&ep=signature&ismsaljsauthenabled__;!!Mih3wA!AIhiGfU3uPoKEyYZGULe3B4YLszx9mBhQGd07e2ZK9SnCFUO7t1H_VXRrWWbSCO6Sxru1LMboIkoQD5astO1_g$>
> > >
> > ________________________________
> > From: eeglablist <eeglablist-bounces at sccn.ucsd.edu> on behalf of Scott
> > Makeig via eeglablist <eeglablist at sccn.ucsd.edu>
> > Sent: Tuesday, May 12, 2026 10:32 AM
> > To: Евгений Машеров <emasherov at yandex.ru>
> > Cc: eeglablist at sccn.ucsd.edu <eeglablist at sccn.ucsd.edu>
> > Subject: Re: [Eeglablist] Persistent 20 Hz Harmonic Line Noise in EEG
> Data
> >
> > [This email originated from outside of UNCW]
> >
> > Eugen,
> >
> > Yes, thanks for correcting me - I was indeed reading from the tutorial
> > example(@ 25 Hz)  rather than the actual data (@ 20 Hz) ... I believe you
> > are correct that some nearby electronic source must have been the cause.
> I
> > believe it would be worthwhile to apply an initial ICA to see whether its
> > scalp projection (scalp map) is stable enough to be largely (or totally)
> > accounted by a single independent component, which then could be removed
> > neatly from the data -- or ignored in further analysis of the (brain)
> > independent source distribution.
> >
> > Scott
> >
> > On Mon, May 11, 2026 at 3:19 PM Евгений Машеров <emasherov at yandex.ru>
> > wrote:
> >
> > > If you're referring to this spectrum,
> > >
> >
> https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Furldefense.com*2Fv3*2F__https*3A*2F*2Fnam12.safelinks.protection.outlook.com*2F*3Furl*3Dhttps*3A*2F*2Furldefense.com*2Fv3*2F__https*3A*2F*2Fibb.co*2F5WbNqYCW__*3B!!Mih3wA!CULZcfFkieH_bUE1w0wwnlMTADY94DKwrrOAJL4cnmw9hjrWFGypYhZLJUl1ZVoht5v4sXALi-8AvrBSXjma*24*26data*3D05*7C02*7Crawlse*40uncw.edu*7C5687d65a15aa4e64d03908deb033b1dc*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639141933252982464*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C*26sdata*3DsYXFOzhWyOH2ACDQyJCVC3IVpqjKiAJvcXasU*2F9nytw*3D*26reserved*3D0__*3BJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSU!!Mih3wA!AIhiGfU3uPoKEyYZGULe3B4YLszx9mBhQGd07e2ZK9SnCFUO7t1H_VXRrWWbSCO6Sxru1LMboIkoQD6kn9dfOw*24&data=05*7C02*7Crawlse*40uncw.edu*7Ce475fe096e8844b757b208deb1cc3912*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639143687887892886*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=*2BVQn*2BUKddwutZUsZsa*2B3uyEpOAX25l9cGdtztz0TTtM*3D&reserved=0__;JSUlJSUlJSUlJSUqKioqKioqKioqKiUlKioqKioqKioqKioqKioqKiUlKiolJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSU!!Mih3wA!AecfBxaMRbnkXpSJ4rvweP9891s7UjaDa-pJhM1RcUuiQnhGpaID0zFJ0jPVfcWaL8UfyBcfEsF2hxpXeaH2cSz-qg$ 
> <https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Furldefense.com*2Fv3*2F__https*3A*2F*2Fibb.co*2F5WbNqYCW__*3B!!Mih3wA!CULZcfFkieH_bUE1w0wwnlMTADY94DKwrrOAJL4cnmw9hjrWFGypYhZLJUl1ZVoht5v4sXALi-8AvrBSXjma*24&data=05*7C02*7Crawlse*40uncw.edu*7C5687d65a15aa4e64d03908deb033b1dc*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639141933252982464*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=sYXFOzhWyOH2ACDQyJCVC3IVpqjKiAJvcXasU*2F9nytw*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSU!!Mih3wA!AIhiGfU3uPoKEyYZGULe3B4YLszx9mBhQGd07e2ZK9SnCFUO7t1H_VXRrWWbSCO6Sxru1LMboIkoQD6kn9dfOw$>
> > <
> >
> https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Furldefense.com*2Fv3*2F__https*3A*2F*2Fibb.co*2F5WbNqYCW__*3B!!Mih3wA!CULZcfFkieH_bUE1w0wwnlMTADY94DKwrrOAJL4cnmw9hjrWFGypYhZLJUl1ZVoht5v4sXALi-8AvrBSXjma*24&data=05*7C02*7Crawlse*40uncw.edu*7Ce475fe096e8844b757b208deb1cc3912*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639143687887915006*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=8OEcXVw7nQyuetqwOTGQNdM3lFusSVuRpFLOeC1gw*2BU*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSU!!Mih3wA!AecfBxaMRbnkXpSJ4rvweP9891s7UjaDa-pJhM1RcUuiQnhGpaID0zFJ0jPVfcWaL8UfyBcfEsF2hxpXeaEBQrg20A$ 
> <https://urldefense.com/v3/__https://ibb.co/5WbNqYCW__;!!Mih3wA!CULZcfFkieH_bUE1w0wwnlMTADY94DKwrrOAJL4cnmw9hjrWFGypYhZLJUl1ZVoht5v4sXALi-8AvrBSXjma$>
> > >
> > > I see a peak at 20 Hz (around 30 dB), 40 Hz (around 40 dB), and 60 Hz
> > > (around 30 dB). Possible sources include MRI (pulse frequency) or
> > > physiotherapy equipment. It could also be some industrial equipment. In
> > one
> > > case, the interference was from a nearby airport's radar.
> > >
> > > > The plots here clearly show the noise is at 25 Hz and its harmonics
> ...
> > > >
> > > > On Sat, May 9, 2026 at 11:08 AM Евгений Машеров via eeglablist <
> > > eeglablist at sccn.ucsd.edu> wrote:
> > > >
> > > >> This isn't power line interference. It would be 50 or 60 Hz, and
> > > possibly harmonics at multiples of 100-150-200... or 120-180-240-...
> > > frequencies. Subharmonics wouldn't appear. It's undoubtedly a technical
> > > artifact, although I'm unsure of the source. Railways sometimes use
> other
> > > frequencies (16 2/3 Hz, others—20, 25, 45, and others no longer in
> > use)—but
> > > not 20 Hz. A television screen will show 50 (PAL/SECAM) or 60 (NTSC)
> Hz,
> > > computer monitors even higher, and it could be some obsolete or
> > specialized
> > > video equipment (with a lower frame rate). I'd assume the source in
> your
> > > case is medical equipment (although I can't rule out, for example, a
> > > security alarm).
> > > >>
> > > >> The best course of action is to find the source of the interference
> > and
> > > turn it off. However, this isn't always possible, especially with
> medical
> > > equipment. A similar situation occurred with an infusion pump in the
> > > intensive care unit: the intermittently activated drive solenoid
> created
> > > regular pulsed interference. A possible solution, also used to combat a
> > > physiological artifact—an electrocardiographic signal interfering with
> > the
> > > EEG—is to evaluate the shape of the interference (the presence of
> > harmonics
> > > indicates that the shape is clearly not sinusoidal) and subtract it
> from
> > > the signal. To do this, narrowband filtering isolates the fundamental
> 20
> > Hz
> > > harmonic, its peaks (or zero crossings) are used as a trigger for
> > > averaging, the averaged response is calculated for each channel, and
> then
> > > subtracted from the signals for the individual channels. Naturally,
> this
> > is
> > > a post-processing method and is not suitable for online processing.
> > > >>
> > > >> Your truly
> > > >>
> > > >> Eugen Masherov
> > > >>
> > > >>> Dear EEGLAB List,
> > > >>
> > > >>>
> > > >>
> > > >>> I am currently dealing with persistent line-noise contamination in
> my
> > > EEG
> > > >>
> > > >>> recordings, specifically strong peaks around 20 Hz and its
> harmonics
> > > (40
> > > >>
> > > >>> Hz, 60 Hz, etc.). The contamination remains substantial even after
> > > multiple
> > > >>
> > > >>> preprocessing attempts.
> > > >>
> > > >>>
> > > >>
> > > >>> Raw Data PSD for reference:
> > >
> >
> https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Furldefense.com*2Fv3*2F__https*3A*2F*2Fnam12.safelinks.protection.outlook.com*2F*3Furl*3Dhttps*3A*2F*2Furldefense.com*2Fv3*2F__https*3A*2F*2Fibb.co*2F5WbNqYCW__*3B!!Mih3wA!EfdhaFV_tpprT4Wc8_HcfK7KDVTXA0O0NPLW3Ws732nxP-TmPxHUXI_qXevcPIO8hfndJk8euI6BWbB1_dCuBxHG_A*24*26data*3D05*7C02*7Crawlse*40uncw.edu*7C5687d65a15aa4e64d03908deb033b1dc*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639141933253242409*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C*26sdata*3DeIaQQjDBsZ5HLbBJlpYPu2JekmgTwMZu5OqKhVX6CsA*3D*26reserved*3D0__*3BJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!AIhiGfU3uPoKEyYZGULe3B4YLszx9mBhQGd07e2ZK9SnCFUO7t1H_VXRrWWbSCO6Sxru1LMboIkoQD7L52136w*24&data=05*7C02*7Crawlse*40uncw.edu*7Ce475fe096e8844b757b208deb1cc3912*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639143687887935575*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=OaZRSXAHx8l0rTTcnmr889ygH9jlQhJkQGzewdXSDd8*3D&reserved=0__;JSUlJSUlJSUlJSUqKioqKioqKioqKiUlKioqKioqKioqKioqKioqKiUlKiUlJSUlJSUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!AecfBxaMRbnkXpSJ4rvweP9891s7UjaDa-pJhM1RcUuiQnhGpaID0zFJ0jPVfcWaL8UfyBcfEsF2hxpXeaFmJmv6dA$ 
> <https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Furldefense.com*2Fv3*2F__https*3A*2F*2Fibb.co*2F5WbNqYCW__*3B!!Mih3wA!EfdhaFV_tpprT4Wc8_HcfK7KDVTXA0O0NPLW3Ws732nxP-TmPxHUXI_qXevcPIO8hfndJk8euI6BWbB1_dCuBxHG_A*24&data=05*7C02*7Crawlse*40uncw.edu*7C5687d65a15aa4e64d03908deb033b1dc*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639141933253242409*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=eIaQQjDBsZ5HLbBJlpYPu2JekmgTwMZu5OqKhVX6CsA*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!AIhiGfU3uPoKEyYZGULe3B4YLszx9mBhQGd07e2ZK9SnCFUO7t1H_VXRrWWbSCO6Sxru1LMboIkoQD7L52136w$>
> > <
> >
> https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Furldefense.com*2Fv3*2F__https*3A*2F*2Fibb.co*2F5WbNqYCW__*3B!!Mih3wA!EfdhaFV_tpprT4Wc8_HcfK7KDVTXA0O0NPLW3Ws732nxP-TmPxHUXI_qXevcPIO8hfndJk8euI6BWbB1_dCuBxHG_A*24&data=05*7C02*7Crawlse*40uncw.edu*7Ce475fe096e8844b757b208deb1cc3912*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639143687887955514*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=JAOGdcNpjN4CyhvhUXJAvq10ZyrvUZMxkt4*2FBD3NO70*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSU!!Mih3wA!AecfBxaMRbnkXpSJ4rvweP9891s7UjaDa-pJhM1RcUuiQnhGpaID0zFJ0jPVfcWaL8UfyBcfEsF2hxpXeaHsVJKGAg$ 
> <https://urldefense.com/v3/__https://ibb.co/5WbNqYCW__;!!Mih3wA!EfdhaFV_tpprT4Wc8_HcfK7KDVTXA0O0NPLW3Ws732nxP-TmPxHUXI_qXevcPIO8hfndJk8euI6BWbB1_dCuBxHG_A$>
> > >
> > > >>
> > > >>>
> > > >>
> > > >>> I have already tried:
> > > >>
> > > >>>
> > > >>
> > > >>> -
> > > >>
> > > >>>
> > > >>
> > > >>> Average re-referencing before notch filtering
> > > >>
> > > >>> -
> > > >>
> > > >>>
> > > >>
> > > >>> Average re-referencing after notch filtering
> > > >>
> > > >>> -
> > > >>
> > > >>>
> > > >>
> > > >>> Multiple full-rank average reference implementations, including:
> > > >>
> > > >>> -
> > > >>
> > > >>>
> > > >>
> > > >>> Makoto’s fullRankAveRef()
> > > >>
> > > >>> -
> > > >>
> > > >>>
> > > >>
> > > >>> Manual full-rank average reference using (nchan + 1) denominator
> > > >>
> > > >>>
> > > >>
> > > >>> Despite this, a large amount of the harmonic noise is still present
> > in
> > > the
> > > >>
> > > >>> spectra.
> > > >>
> > > >>>
> > > >>
> > > >>> For reference, the issue looks somewhat similar to this discussion:
> > > >>
> > > >>>
> > >
> >
> https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Furldefense.com*2Fv3*2F__https*3A*2F*2Fnam12.safelinks.protection.outlook.com*2F*3Furl*3Dhttps*3A*2F*2Furldefense.com*2Fv3*2F__https*3A*2F*2Fmne.discourse.group*2Ft*2Fline-noise-eeg*2F11817__*3B!!Mih3wA!EfdhaFV_tpprT4Wc8_HcfK7KDVTXA0O0NPLW3Ws732nxP-TmPxHUXI_qXevcPIO8hfndJk8euI6BWbB1_dB2ZPVDPg*24*26data*3D05*7C02*7Crawlse*40uncw.edu*7C5687d65a15aa4e64d03908deb033b1dc*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639141933253274372*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C*26sdata*3DFJlsgfiAAdK1fqkeYAwd1vB8OyNv5OSzHhPdOJQVKAI*3D*26reserved*3D0__*3BJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUl!!Mih3wA!AIhiGfU3uPoKEyYZGULe3B4YLszx9mBhQGd07e2ZK9SnCFUO7t1H_VXRrWWbSCO6Sxru1LMboIkoQD4f1vGU-Q*24&data=05*7C02*7Crawlse*40uncw.edu*7Ce475fe096e8844b757b208deb1cc3912*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639143687887974956*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=YREKXSVB*2F1cdeK1tXhUhJkse8suE524Be9hFCGgqrtU*3D&reserved=0__;JSUlJSUlJSUlJSUqKioqKioqKioqKioqJSUqKioqKioqKioqKioqKioqJSUqJSUlJSUlJSUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!AecfBxaMRbnkXpSJ4rvweP9891s7UjaDa-pJhM1RcUuiQnhGpaID0zFJ0jPVfcWaL8UfyBcfEsF2hxpXeaGGW4LeCg$ 
> <https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Furldefense.com*2Fv3*2F__https*3A*2F*2Fmne.discourse.group*2Ft*2Fline-noise-eeg*2F11817__*3B!!Mih3wA!EfdhaFV_tpprT4Wc8_HcfK7KDVTXA0O0NPLW3Ws732nxP-TmPxHUXI_qXevcPIO8hfndJk8euI6BWbB1_dB2ZPVDPg*24&data=05*7C02*7Crawlse*40uncw.edu*7C5687d65a15aa4e64d03908deb033b1dc*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639141933253274372*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=FJlsgfiAAdK1fqkeYAwd1vB8OyNv5OSzHhPdOJQVKAI*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUl!!Mih3wA!AIhiGfU3uPoKEyYZGULe3B4YLszx9mBhQGd07e2ZK9SnCFUO7t1H_VXRrWWbSCO6Sxru1LMboIkoQD4f1vGU-Q$>
> > <
> >
> https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Furldefense.com*2Fv3*2F__https*3A*2F*2Fmne.discourse.group*2Ft*2Fline-noise-eeg*2F11817__*3B!!Mih3wA!EfdhaFV_tpprT4Wc8_HcfK7KDVTXA0O0NPLW3Ws732nxP-TmPxHUXI_qXevcPIO8hfndJk8euI6BWbB1_dB2ZPVDPg*24&data=05*7C02*7Crawlse*40uncw.edu*7Ce475fe096e8844b757b208deb1cc3912*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639143687887995192*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=BQNETjJf*2FjpjQQFo5R5OI2fwUjfUlRjj1n5G32aMjJc*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!AecfBxaMRbnkXpSJ4rvweP9891s7UjaDa-pJhM1RcUuiQnhGpaID0zFJ0jPVfcWaL8UfyBcfEsF2hxpXeaEhgAX_lw$ 
> <https://urldefense.com/v3/__https://mne.discourse.group/t/line-noise-eeg/11817__;!!Mih3wA!EfdhaFV_tpprT4Wc8_HcfK7KDVTXA0O0NPLW3Ws732nxP-TmPxHUXI_qXevcPIO8hfndJk8euI6BWbB1_dB2ZPVDPg$>
> > >
> > > >>
> > > >>>
> > > >>
> > > >>> I would greatly appreciate advice on how people in the EEGLAB
> > community
> > > >>
> > > >>> typically handle this kind of persistent harmonic contamination.
> > > >>
> > > >>>
> > > >>
> > > >>> Would you recommend CleanLine, Zapline-plus, spectrum
> interpolation,
> > > ASR,
> > > >>
> > > >>> ICA-based approaches, or something else?
> > > >>
> > > >>>
> > > >>
> > > >>> Thank you very much for your help.
> > > >>
> > > >>>
> > > >>
> > > >>> Best regards,
> > > >>
> > > >>> Maeghal Jain
> > > >>
> > > >>> BISE Lab
> > > >>
> > > >>> _______________________________________________
> > > >>
> > > >>> To unsubscribe, send an empty email to
> > > eeglablist-unsubscribe at sccn.ucsd.edu or visit
> > >
> >
> https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Furldefense.com*2Fv3*2F__https*3A*2F*2Fnam12.safelinks.protection.outlook.com*2F*3Furl*3Dhttps*3A*2F*2Fsccn.ucsd.edu*2Fmailman*2Flistinfo*2Feeglablist*26data*3D05*7C02*7Crawlse*40uncw.edu*7C5687d65a15aa4e64d03908deb033b1dc*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639141933253298283*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C*26sdata*3DlisqxIA*2B4OVEAiMrGK6cWHUvIUSghekFoxf6DSQHCm8*3D*26reserved*3D0__*3BJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUl!!Mih3wA!AIhiGfU3uPoKEyYZGULe3B4YLszx9mBhQGd07e2ZK9SnCFUO7t1H_VXRrWWbSCO6Sxru1LMboIkoQD6inTlV8w*24&data=05*7C02*7Crawlse*40uncw.edu*7Ce475fe096e8844b757b208deb1cc3912*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639143687888017627*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=aC4WiGaVbhTO6pUKScVqEah1FuXspvgl*2BxQHXqbdNEU*3D&reserved=0__;JSUlJSUlJSUlJSUqKioqKiolJSoqKioqKioqKioqKioqKiolJSoqJSUlJSUlJSUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!AecfBxaMRbnkXpSJ4rvweP9891s7UjaDa-pJhM1RcUuiQnhGpaID0zFJ0jPVfcWaL8UfyBcfEsF2hxpXeaENR3W6VA$ 
> <https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Fsccn.ucsd.edu*2Fmailman*2Flistinfo*2Feeglablist&data=05*7C02*7Crawlse*40uncw.edu*7C5687d65a15aa4e64d03908deb033b1dc*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639141933253298283*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=lisqxIA*2B4OVEAiMrGK6cWHUvIUSghekFoxf6DSQHCm8*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUl!!Mih3wA!AIhiGfU3uPoKEyYZGULe3B4YLszx9mBhQGd07e2ZK9SnCFUO7t1H_VXRrWWbSCO6Sxru1LMboIkoQD6inTlV8w$>
> > <
> https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Fsccn.ucsd.edu*2Fmailman*2Flistinfo*2Feeglablist&data=05*7C02*7Crawlse*40uncw.edu*7Ce475fe096e8844b757b208deb1cc3912*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639143687888037902*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=PfDaMpuE*2FrjkV*2B5Zvh8x7WXgQ6f1v8KZUjH4hjAPcf4*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!AecfBxaMRbnkXpSJ4rvweP9891s7UjaDa-pJhM1RcUuiQnhGpaID0zFJ0jPVfcWaL8UfyBcfEsF2hxpXeaGxh_EnLg$ 
> >  .
> > > >>
> > > >> _______________________________________________
> > > >>
> > > >> To unsubscribe, send an empty email to
> > > eeglablist-unsubscribe at sccn.ucsd.edu or visit
> > >
> >
> https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Furldefense.com*2Fv3*2F__https*3A*2F*2Fnam12.safelinks.protection.outlook.com*2F*3Furl*3Dhttps*3A*2F*2Fsccn.ucsd.edu*2Fmailman*2Flistinfo*2Feeglablist*26data*3D05*7C02*7Crawlse*40uncw.edu*7C5687d65a15aa4e64d03908deb033b1dc*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639141933253319451*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C*26sdata*3D0ZWknYHCiIXcVrW2*2FTmrgY2hufRp4KeeR81*2BCOtUB0w*3D*26reserved*3D0__*3BJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!AIhiGfU3uPoKEyYZGULe3B4YLszx9mBhQGd07e2ZK9SnCFUO7t1H_VXRrWWbSCO6Sxru1LMboIkoQD7Ur3gyKA*24&data=05*7C02*7Crawlse*40uncw.edu*7Ce475fe096e8844b757b208deb1cc3912*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639143687888057540*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=HRsTP8oWIh9JKE*2B4vkKH*2BDonFQPfIEEnoVqGcJQeOqA*3D&reserved=0__;JSUlJSUlJSUlJSUqKioqKiolJSoqKioqKioqKioqKioqKiolJSoqKiUlJSUlJSUlJSUlJSUlJSUlJSUlJSUl!!Mih3wA!AecfBxaMRbnkXpSJ4rvweP9891s7UjaDa-pJhM1RcUuiQnhGpaID0zFJ0jPVfcWaL8UfyBcfEsF2hxpXeaE6_MbJkw$ 
> <https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Fsccn.ucsd.edu*2Fmailman*2Flistinfo*2Feeglablist&data=05*7C02*7Crawlse*40uncw.edu*7C5687d65a15aa4e64d03908deb033b1dc*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639141933253319451*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=0ZWknYHCiIXcVrW2*2FTmrgY2hufRp4KeeR81*2BCOtUB0w*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!AIhiGfU3uPoKEyYZGULe3B4YLszx9mBhQGd07e2ZK9SnCFUO7t1H_VXRrWWbSCO6Sxru1LMboIkoQD7Ur3gyKA$>
> > <
> https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Fsccn.ucsd.edu*2Fmailman*2Flistinfo*2Feeglablist&data=05*7C02*7Crawlse*40uncw.edu*7Ce475fe096e8844b757b208deb1cc3912*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639143687888233110*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=W*2BIzP3Th0SqSHj2P4L8ouazqghJ52hZsWwUcGZ0pj0Q*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUl!!Mih3wA!AecfBxaMRbnkXpSJ4rvweP9891s7UjaDa-pJhM1RcUuiQnhGpaID0zFJ0jPVfcWaL8UfyBcfEsF2hxpXeaFEz-jR_g$ 
> > .
> > > >
> > > > --
> > > > Scott Makeig, Research Scientist and Director, Swartz Center for
> > > Computational Neuroscience, Institute for Neural Computation,
> University
> > of
> > > California San Diego, La Jolla CA 92093-0559,
> >
> https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Furldefense.com*2Fv3*2F__https*3A*2F*2Fnam12.safelinks.protection.outlook.com*2F*3Furl*3Dhttp*3A*2F*2Fsccn.ucsd.edu*2F*scott*26data*3D05*7C02*7Crawlse*40uncw.edu*7C5687d65a15aa4e64d03908deb033b1dc*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639141933253340446*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C*26sdata*3D0ireFT9mF795bN4bgEOg9SzzOiBzfbLhvBEBSl*2FXZKg*3D*26reserved*3D0__*3BJSUlJX4lJSUlJSUlJSUlJSUlJSUlJSU!!Mih3wA!AIhiGfU3uPoKEyYZGULe3B4YLszx9mBhQGd07e2ZK9SnCFUO7t1H_VXRrWWbSCO6Sxru1LMboIkoQD7giQR85g*24&data=05*7C02*7Crawlse*40uncw.edu*7Ce475fe096e8844b757b208deb1cc3912*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639143687888267081*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=nxA*2BBOBJpzUBRjbTYEH04p3KuqqUWRRY5S*2FS5fUisUM*3D&reserved=0__;JSUlJSUlJSUlJSUqKioqKiUlKioqKioqKioqKioqKioqKiUlKiolJSUlJSUlJSUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!AecfBxaMRbnkXpSJ4rvweP9891s7UjaDa-pJhM1RcUuiQnhGpaID0zFJ0jPVfcWaL8UfyBcfEsF2hxpXeaH8QouWCg$ 
> <https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=http*3A*2F*2Fsccn.ucsd.edu*2F*scott&data=05*7C02*7Crawlse*40uncw.edu*7C5687d65a15aa4e64d03908deb033b1dc*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639141933253340446*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=0ireFT9mF795bN4bgEOg9SzzOiBzfbLhvBEBSl*2FXZKg*3D&reserved=0__;JSUlJX4lJSUlJSUlJSUlJSUlJSUlJSU!!Mih3wA!AIhiGfU3uPoKEyYZGULe3B4YLszx9mBhQGd07e2ZK9SnCFUO7t1H_VXRrWWbSCO6Sxru1LMboIkoQD7giQR85g$>
> > <
> https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=http*3A*2F*2Fsccn.ucsd.edu*2F*scott&data=05*7C02*7Crawlse*40uncw.edu*7Ce475fe096e8844b757b208deb1cc3912*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639143687888288255*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=jF8oOxdxJIP*2BRsEo8LLLvMV0qI*2FZ9G9B10KUXU81o3A*3D&reserved=0__;JSUlJX4lJSUlJSUlJSUlJSUlJSUlJSUl!!Mih3wA!AecfBxaMRbnkXpSJ4rvweP9891s7UjaDa-pJhM1RcUuiQnhGpaID0zFJ0jPVfcWaL8UfyBcfEsF2hxpXeaGYcPzuKA$ 
> >
> > >
> >
> >
> > --
> > Scott Makeig, Research Scientist and Director, Swartz Center for
> > Computational Neuroscience, Institute for Neural Computation, University
> of
> > California San Diego, La Jolla CA 92093-0559,
> >
> https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Furldefense.com*2Fv3*2F__https*3A*2F*2Fnam12.safelinks.protection.outlook.com*2F*3Furl*3Dhttp*3A*2F*2Fsccn.ucsd.edu*2F*scott*26data*3D05*7C02*7Crawlse*40uncw.edu*7C5687d65a15aa4e64d03908deb033b1dc*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639141933253360104*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C*26sdata*3DAexTc5v9lgdNZw1jdpPGQgxXlcV6MHNOsEemQodk5bs*3D*26reserved*3D0__*3BJSUlJX4lJSUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!AIhiGfU3uPoKEyYZGULe3B4YLszx9mBhQGd07e2ZK9SnCFUO7t1H_VXRrWWbSCO6Sxru1LMboIkoQD7fiWqVIw*24&data=05*7C02*7Crawlse*40uncw.edu*7Ce475fe096e8844b757b208deb1cc3912*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639143687888308865*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=Tr8yDZWxy5oJ3Ilhc9*2FUIhr*2FMAzGk9CF00Y4gTdlbqo*3D&reserved=0__;JSUlJSUlJSUlJSUqKioqKiUlKioqKioqKioqKioqKioqKiUlKiUlJSUlJSUlJSUlJSUlJSUlJSUlJSUl!!Mih3wA!AecfBxaMRbnkXpSJ4rvweP9891s7UjaDa-pJhM1RcUuiQnhGpaID0zFJ0jPVfcWaL8UfyBcfEsF2hxpXeaGvlSp_4g$ 
> <https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=http*3A*2F*2Fsccn.ucsd.edu*2F*scott&data=05*7C02*7Crawlse*40uncw.edu*7C5687d65a15aa4e64d03908deb033b1dc*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639141933253360104*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=AexTc5v9lgdNZw1jdpPGQgxXlcV6MHNOsEemQodk5bs*3D&reserved=0__;JSUlJX4lJSUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!AIhiGfU3uPoKEyYZGULe3B4YLszx9mBhQGd07e2ZK9SnCFUO7t1H_VXRrWWbSCO6Sxru1LMboIkoQD7fiWqVIw$>
> > <
> https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=http*3A*2F*2Fsccn.ucsd.edu*2F*scott&data=05*7C02*7Crawlse*40uncw.edu*7Ce475fe096e8844b757b208deb1cc3912*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639143687888333408*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=fVZx5vVCmiQJmaTltkxQuJdaIdHK3jCoaLET0TPsge8*3D&reserved=0__;JSUlJX4lJSUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!AecfBxaMRbnkXpSJ4rvweP9891s7UjaDa-pJhM1RcUuiQnhGpaID0zFJ0jPVfcWaL8UfyBcfEsF2hxpXeaHc-a-y_Q$ 
> >
> > _______________________________________________
> > To unsubscribe, send an empty email to
> > eeglablist-unsubscribe at sccn.ucsd.edu or visit
> >
> https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Furldefense.com*2Fv3*2F__https*3A*2F*2Fnam12.safelinks.protection.outlook.com*2F*3Furl*3Dhttps*3A*2F*2Fsccn.ucsd.edu*2Fmailman*2Flistinfo*2Feeglablist*26data*3D05*7C02*7Crawlse*40uncw.edu*7C5687d65a15aa4e64d03908deb033b1dc*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639141933253379386*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C*26sdata*3DrylAi2wjbaU*2FPmfgMeLbNrbHSVWdvNxNV62U*2FMZmkrQ*3D*26reserved*3D0__*3BJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!AIhiGfU3uPoKEyYZGULe3B4YLszx9mBhQGd07e2ZK9SnCFUO7t1H_VXRrWWbSCO6Sxru1LMboIkoQD796RxNIg*24&data=05*7C02*7Crawlse*40uncw.edu*7Ce475fe096e8844b757b208deb1cc3912*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639143687888352393*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=S*2BQ6mZdGQmP3o3Jnw*2F5SiSd6voMdaFoYojtgY3fS7LE*3D&reserved=0__;JSUlJSUlJSUlJSUqKioqKiolJSoqKioqKioqKioqKioqKiolJSoqKiUlJSUlJSUlJSUlJSUlJSUlJSUlJSUl!!Mih3wA!AecfBxaMRbnkXpSJ4rvweP9891s7UjaDa-pJhM1RcUuiQnhGpaID0zFJ0jPVfcWaL8UfyBcfEsF2hxpXeaF_ihBkSw$ 
> <https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Fsccn.ucsd.edu*2Fmailman*2Flistinfo*2Feeglablist&data=05*7C02*7Crawlse*40uncw.edu*7C5687d65a15aa4e64d03908deb033b1dc*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639141933253379386*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=rylAi2wjbaU*2FPmfgMeLbNrbHSVWdvNxNV62U*2FMZmkrQ*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!AIhiGfU3uPoKEyYZGULe3B4YLszx9mBhQGd07e2ZK9SnCFUO7t1H_VXRrWWbSCO6Sxru1LMboIkoQD796RxNIg$>
> > <
> https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Fsccn.ucsd.edu*2Fmailman*2Flistinfo*2Feeglablist&data=05*7C02*7Crawlse*40uncw.edu*7Ce475fe096e8844b757b208deb1cc3912*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639143687888374091*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=4nXXdpKNDap*2FY45Ltw3lhNGe011w50GAE*2B31dc3F0y8*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJQ!!Mih3wA!AecfBxaMRbnkXpSJ4rvweP9891s7UjaDa-pJhM1RcUuiQnhGpaID0zFJ0jPVfcWaL8UfyBcfEsF2hxpXeaFjKqlTHA$ 
> >.
> > _______________________________________________
> > To unsubscribe, send an empty email to
> > eeglablist-unsubscribe at sccn.ucsd.edu or visit
> >
> https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Fsccn.ucsd.edu*2Fmailman*2Flistinfo*2Feeglablist&data=05*7C02*7Crawlse*40uncw.edu*7Ce475fe096e8844b757b208deb1cc3912*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639143687888394167*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=fH*2Bw3Kbc5rOGwhaKAFd*2Fsq7Ebx0vmsNWK*2BIu9z*2BL5Q0*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUl!!Mih3wA!AecfBxaMRbnkXpSJ4rvweP9891s7UjaDa-pJhM1RcUuiQnhGpaID0zFJ0jPVfcWaL8UfyBcfEsF2hxpXeaGvV1Cmmw$ 
> <https://sccn.ucsd.edu/mailman/listinfo/eeglablist >  .
> > _______________________________________________
> > To unsubscribe, send an empty email to
> > eeglablist-unsubscribe at sccn.ucsd.edu or visit
> >
> https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Fsccn.ucsd.edu*2Fmailman*2Flistinfo*2Feeglablist&data=05*7C02*7Crawlse*40uncw.edu*7Ce475fe096e8844b757b208deb1cc3912*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639143687888414442*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=XQrFBkFtn5pd25mRxJF3*2BvCnBw7pNPfVZerJZTqXWUE*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUl!!Mih3wA!AecfBxaMRbnkXpSJ4rvweP9891s7UjaDa-pJhM1RcUuiQnhGpaID0zFJ0jPVfcWaL8UfyBcfEsF2hxpXeaHaj4eByQ$ 
> <https://sccn.ucsd.edu/mailman/listinfo/eeglablist > .
> _______________________________________________
> To unsubscribe, send an empty email to
> eeglablist-unsubscribe at sccn.ucsd.edu or visit
> https://urldefense.com/v3/__https://nam12.safelinks.protection.outlook.com/?url=https*3A*2F*2Fsccn.ucsd.edu*2Fmailman*2Flistinfo*2Feeglablist&data=05*7C02*7Crawlse*40uncw.edu*7Ce475fe096e8844b757b208deb1cc3912*7C2213678197534c75af2868a078871ebf*7C0*7C0*7C639143687888433918*7CUnknown*7CTWFpbGZsb3d8eyJFbXB0eU1hcGkiOnRydWUsIlYiOiIwLjAuMDAwMCIsIlAiOiJXaW4zMiIsIkFOIjoiTWFpbCIsIldUIjoyfQ*3D*3D*7C0*7C*7C*7C&sdata=wtaGDZi*2By1kePAE4NmxRay1t3gl8hsz7EsLM55f5BxE*3D&reserved=0__;JSUlJSUlJSUlJSUlJSUlJSUlJSUlJSUl!!Mih3wA!AecfBxaMRbnkXpSJ4rvweP9891s7UjaDa-pJhM1RcUuiQnhGpaID0zFJ0jPVfcWaL8UfyBcfEsF2hxpXeaGi2BeLag$ 
> <https://sccn.ucsd.edu/mailman/listinfo/eeglablist >.
>



More information about the eeglablist mailing list